[time-nuts] 5 MHZ PIC PPS Divider?
Bruce Griffiths
bruce.griffiths at xtra.co.nz
Thu Apr 17 01:14:56 UTC 2008
Brooke Clarke wrote:
> Hi Bruce:
>
> The A/D used is only 12 bits @ 300 kHz and Linear now has very similar 12 bit
> A/D at 600 kHz and 24 bit much slower. Since a s/h is used to capture the ramp
> value the key spec seems to be that the hold time exceeds the conversion time.
> Are there s/h circuits that hold for more than 130 ms?
>
> Hard to say what "hold" means when talking about 24 bit conversions. I'm not
> sure of how droop would effect the result.
>
> I'm asking because the LTC2400 is about $8 compared to the LTC1273A for $25.
>
> Have Fun,
>
> Brooke Clarke
>
Brooke
You would need a separate buffered sample and hold when using an LTC2400
or similar sigma delta ADC as opposed to a capacitive charge
redistribuiton ADC like the LTC1273 and its successors.
Perhaps the best way to extend the hold time is to resample the analog
voltage with a slower acquisition time low leakage sample and hold circuit.
Since the output is unipolar one could just ramp up an integrator until
its output is equal to the voltage across the TAC capacitor.
The LTC2400 has a periodic switched capacitor input so that it
approximates a resistive load, thus the TAC output needs to be buffered.
If you use guard rings etc the TAC leakage/discharge current may be
around 100pA or so with a runup current of say 5mA. Thus discharge time
for a 400ns full scale input will be around 20secs.
Droop as long as its linear wont affect conversion linearity however it
will affect the TAC gain.
You would need to use a very low bias current buffer to drive the
LTC2400, however its settling time can be quite slow.
It would also be possible to use an modified integrator with a large
shunt input cap for temporary charge storage whilst the integrator settles.
I suspect it would be better to use a cheaper more modern charge
redistribution ADC like an LTC1415 or similar.
Bruce
More information about the Time-nuts_lists.febo.com
mailing list