[time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?
Prologix
support at prologix.biz
Wed Aug 27 22:31:25 UTC 2008
Bill,
How are you reading from 5334A? Is the read-after-write mode OFF?
I recommend the following sequence:
++auto 0 -- turn off read-after-write
++read eoi -- read until EOI is asserted by the instrument
++read eoi
....
(If your instrument does not assert EOI, specify a character to terminate
++read command. See manual for details.)
Yes, ++read will address the instrument to talk, read any output, and then
address it to listen.
We are testing an update that resets the controller if it detects a hung
GPIB bus. I can send that to you offline.
Regards,
Abdul
-----Original Message-----
From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com] On
Behalf Of wje
Sent: Wednesday, August 27, 2008 3:12 PM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?
I just got one of the new networked GPIB controllers, and I've been
having some issues. I'm not sure if it's the Prologix or the
instruments, or both.
With the 5370A, I can get samples for some period of time ranging from
about 2 to 4 hours before the GPIB controller stops responding. I
suspect this one might be the 5370A -488 controller bug I've heard of.
However, there is no way to recover without power-cycling the Prologix.
Even if this is the 5370A, the Prologix should really have an
asynchronous reset capability.
With the 5334A, I can get samples for as long as I want, no problem.
However, if I set the 5334A to 'WA1' mode, which is supposed to wait
until the counter is addressed before taking each sample, what happens
is that I can read one value. Every successive read returns the same
value; the 5334A never triggers for a new sample. Either I don't
understand the behavior of WA1, although HP's sample program indicates
that what I've described is proper, or there is some incompatibility
between the Prologix and the 5334A.
Unfortunately, the documentation for the Prologix is scanty at best.
There is no description of the actual IEEE-488 handshake process that's
going on. For example, does the Prologix do an unlisten after a read?
The HP docs seem to indicate that in WA1 mode, a sample starts when the
5334A is addressed. So, if it's never unlistened, perhaps it never takes
a new sample?
Any thoughts or suggestions would be appreciated.
--
Bill Ezell
----------
They said 'Windows or better'
so I used Linux.
_______________________________________________
time-nuts mailing list -- time-nuts at febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list