[time-nuts] HP Z3801A Question
ernieperes at aol.com
ernieperes at aol.com
Thu Nov 13 22:37:57 UTC 2008
Hello Said,
Thanks for the detailed reply, it makes sense.
But both unit , the HP and the FURY board running under the same
software simultaneously / GPS Con / and I was mentioning the same small
"windows" TI to GPS, and the FURY displays much better figures then
HP.....but probably each units gives out / display / different
data??????
Also the graphic display / BLUE / on the FURY board the max ampl
between +2,2nsec and -2,0nsec sometimes +0,7nsec ,
-1,7nsec...
the HP unit never goes under +15nsec and -20nsec...... almost 10 times
worse.....
Anyhow I am happy because it seems to me that the HP is still OK after
3 years of sleep.....
Rgsa Ernie.
-----Original Message-----
From: SAIDJACK at aol.com
To: time-nuts at febo.com
Sent: Thu, 13 Nov 2008 10:41 pm
Subject: Re: [time-nuts] HP Z3801A Question
Hello Ernie,
as Bjoern mentioned I think the HP is indicating Time Interval between
GPS
and OCXO 1PPS in nanoseconds, while the Fury is indicating estimated
frequency
accuracy, a unitless number.
The Fury is showing estimated average frequency error over the last
1000
seconds, similar to AVAR 1000s.
You can run GPScon on Fury too, and should be able to compare against
the
graphic you did for the HP unit.
Or you could set the LCD of the Fury to show Time Interval to UTC, or
probe
this time interval to UTC with the command ptim:tint? that will return
the
offset in nanoseconds as well.
An offset of +/-50ns to UTC seems to be normal for the HP unit. On the
Fury,
I would expect this to be +/-10ns on a good OCXO.
To the benefit of the Z3801A unit: if the HP unit indicates a larger
offset,
it may also be because it's internal GPS is not as good as the Fury
GPSDO
receiver. It does not necessarily mean the OCXO is causing the error,
it could
be that the HP GPS is causing the error, and the HP OCXO is still
dead-on
because it filters out its higher GPS errors.
bye,
Said
In a message dated 11/13/2008 02:45:57 Pacific Standard Time,
ernieperes at aol.com writes:
And running parallel the FURY board with a TEMEX OCXO and displays
+1,6E-10 is the average very often 10E-11 and sometimes for a relative
short time / 4-5 min / 1,4E-12.
Any idea why the Z3801A not displaying better figure?? Something wrong
with the HP unit????
Thanks Ernie.