[time-nuts] DMTD to MMTD
Bob Camp
lists at cq.nu
Thu Feb 18 17:41:58 UTC 2010
Hi
Just to bound the problem a little.
An EP3C5 fpga, with all the support stuff (flash chip, regulators..) is
under $30. It's gross overkill for the application.
A set of Mini-Circuits SYPD-2 mixers at $19 each are $114 for six channels.
A set of 20 stage limiters at $2 per op amp would be $240 for the six
channels. They could be $1, they also could be $5 each.
The FPGA is not what leaps out on the bill of material. It's down in the 10%
round off error range....
Bob
-----Original Message-----
From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com] On
Behalf Of Bruce Griffiths
Sent: Thursday, February 18, 2010 7:28 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] DMTD to MMTD
Magnus Danielson wrote:
> Bruce Griffiths wrote:
>> Lux, Jim (337C) wrote:
>>>> -----Original Message-----
>>>> From: time-nuts-bounces at febo.com
>>>> [mailto:time-nuts-bounces at febo.com] On Behalf Of Bruce Griffiths
>>>> Sent: Wednesday, February 17, 2010 2:10 PM
>>>> To: Discussion of precise time and frequency measurement
>>>> Subject: Re: [time-nuts] DMTD to MMTD
>>>>
>>>> The latest version actually records time stamps from a continuously
>>>> running counter clocked at some at a constant frequency (100Mhz??)for
>>>> all channels simultaneously.
>>>> They may use a flag bit for each for each channel to indicate to which
>>>> channel or channels the zero crossing time stamp belongs.
>>> Simpler than that.. it grabs 20 bit numbers and shoves them out in
>>> ASCII over a com port with an indication of which channel it was for.
>>> The FPGA has a 20 bit free running counter at 100 MHz. When an input
>>> changes state, it latches the counter, and shoves it out along with
>>> the channel number. They use an offset frequency>100 Hz so that you
>>> don't have to disambiguate the counter rollovers. (20 bits rolls
>>> over every 10+milliseconds counting at 100 MHz)
>>>
>>> I don't know if there's a FIFO in front of the UART (e.g. what if
>>> you get simultaneous zero crossings).. but I would expect there is.
>>>
>>> The "hard work" is in the zero crossing detector ahead of the FPGA.
>>> (and perhaps in the latching of the ZCD inputs into the FPGA).
>>>
>>> Given how long ago it was made, that FPGA isn't a huge one.
>>>
>> Using 8 flag bits (one per channel) together with the associated time
>> stamp is a little more efficient and very easy to do and it doesn't
>> require a FIFO to ensure that simultaneous zero crossings aren't missed.
>
> It doesn't help at all for this application. The 8 channels is more
> likely to be spread out and as Jim pointed out, the next bin is
> sufficient for needing a unique time-stamp. The flag-solution is less
> efficient (8 bits rather than 3) and the FIFO need is always there,
> but may be implemented in various ways. For the flag system to be
> efficient, a high probability for the same time-bins to be used needs
> to exist, and a high resolution system can expect to actually see
> noise spread out the channels over the time-bins.
>
Depending on the system constraints it may be the difference between
being able to do implement it or not.
> A DMTD systems have a low rate of events per channel, but the nominal
> distance of events for each channel is fairly long, worst-case burst
> is when all channels time-stamps. For a 100 Hz beating and 100 MHz
> clock, the nominal rate of rise/fall events is 200 Hz or 5 ms. Letting
> the locked value stay put for at least 4 ms. If all channels could be
> emptied within these 4 ms (just another way of saying that it has
> enough transport capacity) then a fairly simple schedule system can
> loop through the channels to find a new sample to transmit.
>
> I think the re-occurring flag system should be put to rest, it doesn't
> contribute and is a red herring, at least for this application.
>
Nonsense, it requires simpler logic and for a device with limited
internal connection/routing capability and a large number of channels
the data path interconnections may be simpler and easier to route. It
may also run with a higher clock frequency.
It should even be possible to impement in a relatively small CPLD albeit
with an external FIFO or equivalent (eg a PPI port on a Blackfin DSP).
Each additional channels requires one input pin, one output pin, a 2 bit
synchroniser and a 1 bit wider data path and little else.
> Cheers,
> Magnus
>
Bruce
_______________________________________________
time-nuts mailing list -- time-nuts at febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list