[time-nuts] Advice on 10 MHz isolation/distribution
Garry Thorp
GThorp at pascall.co.uk
Fri Feb 19 15:17:44 UTC 2010
Hello Clay,
Joining in this discussion at a rather late stage - have you
considered using 74AC series gates as buffers? They provide reasonable
isolation and have surprisingly low phase noise.
A single 74AC04 inverter gives over 40dB reverse isolation at 10MHz, so
3 cascaded gates would give more than enough. You would need to use a
separate IC for each stage to achieve the isolation, but they are cheap!
With a series output resistor, a 74AC gate with 5V supply will give
~13dBm while providing a 50R source match. Adding a series tuned circuit
will give a sine wave if required. At 10MHz the first 2 stages will draw
~2mA and the output stage ~15mA (as a matched 50R source), so the
necessary isolation between power supplies can readily be achieved by
using separate RC filter chains to each IC.
I haven't measured the phase noise of an AC04, but I have tried dividing
a low-noise 100MHz OCXO using a 74AC163. At 100MHz, the OCXO's phase
noise was ~-80dBc/Hz at 1Hz offset, -110 at 10Hz, -140 at 100Hz, ~-166
at 1kHz and ~-180 at 10kHz and beyond.
Taking the Qc output of the AC163, the phase noise at 12.5MHz showed the
theoretical 18dB reduction at low offsets, i.e. ~-98 at 1Hz and -128 at
10Hz. It then went into a flicker of phase region, ~-155 at 100Hz and
-165 at 1kHz, reaching a floor of ~-178dBc/Hz by ~100kHz offset.
Extrapolating the flicker region downwards suggests the divider output
phase noise would be ~-145 at 10Hz and -135 at 1Hz. Intuitively, I
wouldn't expect an inverter to have worse phase noise than a counter
from the same family. Cascading 3 inverters would increase the flicker
phase noise by ~5dB, which I think would still be well within the spec
you gave earlier. This approach has the advantage that it can be done
without transformers or inductors (unless you need a sine wave output).
Digital inputs need a high slew rate to achieve low phase noise, so if
your oscillator has a sine wave output you would probably need to square
it up with a (non-saturating) limiter such as a common-base driving into
a Schottky diode limiter, or a long-tailed pair.
As CMOS is a saturating logic family, low-noise power supply is vital.
The input switching threshold is approximately half the supply voltage,
so supply noise + non-infinite slew rate = jitter. The lowest-noise LDO
regulators are probably not good enough. However this is a (relatively!)
straightforward low-frequency problem, that can be solved by using a
heavily filtered voltage reference with a low-noise op amp buffer or
Darlington emitter follower.
Garry
Pascall Electronics Ltd - Registered in England No: 1316674 VAT Registration No: GB 448705134 Registered Office: Brunswick Road, Cobbs Wood, Ashford, Kent, TN23 1EH
The transfer of any controlled technology contained in or attached to this email is covered by the "Open General Export Licence (Technology for Military Goods)", granted by the United Kingdom Secretary of State.
The information contained in this email is provided as a personal communication of the sender and, as such, is not binding on Pascall Electronics Ltd. unless the intended recipient has been notified by signed postal or fax communication that the sender is authorised to commit Pascall Electronics Ltd. on the subject matter concerned.The contents of this email are confidential to the intended recipient at the email address to which it has been addressed. It may not be disclosed to or used by anyone other than this addressee, nor may it be copied in any way. If received in error, please contact Oli Poole, at opoole at pascall.co.uk, quoting the name of the sender and the addressee and then delete it from your system.
More information about the Time-nuts_lists.febo.com