[time-nuts] 5061a now look at wrong freq A1 fault
John Miles
jmiles at pop.net
Sat Feb 27 04:38:14 UTC 2010
I think the best policy re: initial setup of the 5061A is to follow the
manual. There are clear instructions on how to set the beam current and
second harmonic indications and tune the OCXO to find the central peak. If
it is locked but not exactly at 5 MHz, either your C-field is grossly off,
or (more likely) you are tuned to a side peak. But the beam current should
be verified before worrying about that.
-- john, KE5FX
> -----Original Message-----
> From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com]On
> Behalf Of Dott. Alfredo Rosati
> Sent: Friday, February 26, 2010 7:36 PM
> To: J. L. Trantham; time-nuts at febo.com
> Subject: Re: [time-nuts] 5061a now look at wrong freq A1 fault
>
>
> the problem was in the syinthesizer board A1 , the xtal oscillator
> 12.63 was to much aging and was at the limit of pll , now I fixed this
> board , now the 12.63 signal is clear and loocked , before was
> very noisy .
> But seeme there are more problem , the signal now is very close to
> 5MHZ but there are some difference from my rubidium and my Z3805a
> The 5 MHZ now is near 5 x 10-10 .
>
> Now I need to ascertain , if there are more fault , or because I have an
> high performance tube , I need a degausser ??????
>
> I will continue to study .
>
> all suggestion will be appreciated
>
> i5uxj
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
More information about the Time-nuts_lists.febo.com
mailing list