[time-nuts] Achieving maximum performance when driving 5370A/B inputs
Bruce Griffiths
bruce.griffiths at xtra.co.nz
Sun Feb 28 00:01:52 UTC 2010
The attached excerpts from the 5370A and 5370B manuals indicate that for
best performance, that the common practice of driving the 5370A/B 1x
inputs directly from a 5V CMOS logic signal is a bad idea.
For the 5370A attenuating the 5V CMOS signal to a 1V swing with the
threshold set to 0.5V is close to optimum.
An input signal with limits of 0V and +1.4V with a trigger threshold of
0.7V is the maximum usable (for high performance).
An input signal with limits of 0V and +0.3V with a trigger threshold of
0.15V is the minimum usable (for high performance).
For the 5370A attenuating the 5V CMOS signal to a 2V swing with the
threshold set to 1V is close to optimum.
An input signal with limits of 0V and +3.5V with a trigger threshold of
0.7V is the maximum usable (for high performance).
An input signal with limits of 0V and +0.3V with a trigger threshold of
0.15V is the minimum usable (for high performance).
Bruce
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 5370ATriggering.png
Type: image/png
Size: 12169 bytes
Desc: not available
URL: <http://febo.com/pipermail/time-nuts_lists.febo.com/attachments/20100228/c1f3be42/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 5370BTriggering.png
Type: image/png
Size: 64408 bytes
Desc: not available
URL: <http://febo.com/pipermail/time-nuts_lists.febo.com/attachments/20100228/c1f3be42/attachment-0001.png>
More information about the Time-nuts_lists.febo.com
mailing list