[time-nuts] DMTD Terminations II
Bob Camp
lists at rtty.us
Wed Mar 3 12:04:33 UTC 2010
Hi
These days you can get (for a price) NPO caps up above .1 uf. That makes super high value coils less important in a low frequency beat note system. If you are trying to measure phase noise at 100 KHz then 0.1 uf to ground at multiple places likely is not a real good idea.
Bob
On Mar 2, 2010, at 11:28 PM, Brian Kirby wrote:
> I got to correspond with a person at NIST today about DMTD - about the HP 10514B Mixers. They pointed me to a application note at http://www.thegleam.com/ke5fx/stellex/5952-8217.pdf (hey - thats one of our time-nuts!) that shows driving the 10514 mixer at +13 dbm for the LO port and 0 dbm for the RF port.
>
> This application note is a good read. HP recommends terminating the mixer into a 3900 pF cap and 50 ohms in series for the RF termination. ANd they use a LC filter for their VNA set for 90 Khz. I also looked at the schematic of the 8407A were I got the mixers from. They terminate the IF into a 1100 pf cap in series with a 51 ohm resistor, and then it passes thru a 2.7 uH choke going into an amplifier - the IF I beleive is 298 Khz for this unit.
>
> I also did a reconfiguration of my temporary DMTD experiment that has also improved the noise floor. Drawing attached. Also changed the drive levels.
>
> For reference the previous configuration showed 2x10-11 for 0.1 sec; 2x10-12 for 1 sec, 5x10-13 for 10 sec, 6x10-14 for 100 sec, 7x10-15 for 1000 sec. for the noise floor.
>
> The current configuration (of the drawing attached) shows 8x10-12 for 0.1 sec; 8x10-13 for 1 sec, 8x10-14 for 10 sec, 8x10-15 for 100 sec and 8x10-16 for 1000 sec. I used a little more attention in sheilding and spacing to eliminate crosstalk. Note the noise shows some linearity this time....plus more drive....
>
> Brian - KD4FM
>
>
>
>
> <DMTD_Plans_1.1.pdf>_______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list