[time-nuts] HP 5371/2 error
Mark Sims
holrum at hotmail.com
Sun Mar 21 01:37:41 UTC 2010
Basically it means the unit has lost the battery backup to the calibration memory.
The cure is a rather simple calibration step. On the 5372A you input a 60 mV 10KHz sinewave and perform Test 25.
The adjustment is also called "Maximum Hysteresis"