[time-nuts] Making a HP 10811 better
Bob Camp
lists at rtty.us
Mon Mar 29 00:47:42 UTC 2010
Hi
The "best at tha cold end" was a function of where the oven was relative to the turn temperature of the crystal. There's actually a bit more to it than that, but that's the simple answer.
Bottom line - the temperature curve of *your* 10811 likely will be a bit different than what they show.
A cooler, a small bag of ice, a thermometer, and a sunny deck in the summer will give you a temperature ramp. That and some patience will give you a pretty good idea of what your oscillator actually looks like. Other techniques may also work, most provide less of an excuse to drink beer during the experiment.....
Bob
On Mar 28, 2010, at 4:20 PM, Ed Palmer wrote:
> Have you considered putting the 10811 in a deep freeze? :-)
>
> The March 1981 issue of the HP Journal includes an in-depth article on the 10811A/B. It talks about the heater, thermal gain, optimization, etc. Figure 5 on page 21 shows a graph of ambient temperature vs. frequency change. The figure shows that the lowest frequency changes occur around an ambient temperature of -20C! Maybe instead of an outer oven we need to use a Peltier cooler.
>
> http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1981-03.pdf
>
> Ed
>
> WarrenS wrote:
>>
>> Richard wrote approx
>>> ... This is how the Thermal gain can be increase to eliminate temperature drift over typical room variations.
>>
>> Thank You
>> Great Idea. Wish I would of thought of that. But Yeah almost as good, Now I know it, can take that away.
>> That is the type of useful and not obvious information that I was hoping for.
>>
>> THANKS SO MUCH
>> That goes to the top of my list to find a procedure to do it.
>> You are a true Nut hero.
>>
>> WS
>>
>> ********************
>>
>> ----- Original Message ----- From: "Richard H McCorkle" <mccorkle at ptialaska.net>
>> To: "Discussion of precise time and frequency measurement" <time-nuts at febo.com>
>> Sent: Sunday, March 28, 2010 11:32 AM
>> Subject: Re: [time-nuts] Making a HP 10811 better
>>
>>
>>> Warren,
>>> One thing Rick Karlquist pointed out is a higher thermal gain can be
>>> realized to minimize temperature effects by optimizing the heater
>>> transistor balance. The 10811A/B Quartz Crystal Oscillator Operating
>>> & Service Manual describes the balance circuit in section 8-40. The
>>> two heater transistors are not equally spaced with Q7 being closer
>>> to the crystal so rather than 50% of the current flowing thru both
>>> heater transistors the current in Q8 is set slightly higher at 57%
>>> +/- 2% in a production unit to apply equal heat at the crystal.
>>> The typical thermal gain on a production unit is on the order of
>>> 100 without optimization, but gains of 1000 or more are possible if
>>> the heater transistor balance is optimized. I was fortunate enough
>>> to obtain a set of single oven 10811-60158 units that had been
>>> optimized by an HP engineer for different temperature sensitivities
>>> and tagged as such. The unit optimized for maximum thermal gain is
>>> virtually immune to typical room temperature variations. A second
>>> unit optimized for double oven operation has much lower thermal
>>> gain and makes a great unit for testing software temperature
>>> compensation routines and outer oven designs.
>>>
>>> Richard
>>>
>>>>
>>>> Time to Push the "reset button"
>>>>
>>>> I hope we can all agree what one is not going to find an axes that makes the
>>>> Frequency modulation that is caused by tapping on the Oscillator, or the
>>>> table or the airplane, or the boat, to go away, for whatever reasons.
>>>>
>>>> Sorry, what I've been falsely referring to as the zero-G axes is actually
>>>> the "Zero Tilt" angle axes.
>>>> I have not had problems with My G changing short term, (BUT it would be
>>>> interesting to see if I can detect the moon overhead).
>>>> What I do have is some problems with the Osc tilt angle changing due to its
>>>> position changing a little.
>>>> It is the Zero-tilt angle axes that works over a very small change of a few
>>>> degrees at most.
>>>>
>>>> On the other hand, the Zero tilt axes is in fact the Max G sensitivity axes.
>>>> If you want the Min G sensitivity axes, so that there is no change when
>>>> turning the Osc over, That is 90 deg from the zero tilt axes, which is also
>>>> how you make the Osc into its best Tilt angle meter.
>>>> What happens at any G value that is between +1 and -1, or is greater than +1
>>>> or -1, I have not tested for, so I'm not qualified to speculate.
>>>> I am only stating that there is an axes where the Oscillator frequency is
>>>> exactly the same when you turn it exactly over.
>>>> AND this is NOT the axes you want to have, If it is going to be tilted even
>>>> a sub sub fraction of a deg.
>>>>
>>>> Now can we get back to making the 10811 Osc better?
>>>>
>>>> ws
>>>>
>>>> **********************
>>>> *************************
>>>> From: "Bob Camp" <lists at rtty.us>
>>>> Hi
>>>>
>>>> If your zero g axis only works over a 1/100 G range, you are looking at
>>>> something other than acceleration. If you have found an axis with a zero, it
>>>> should be a just as much a null at 1 mG as at 1 G as at 10 G's. It's also
>>>> possible that your "zero" is actually a minimum below your test resolution
>>>> and higher G's bring it up to the point you can measure it. There's no
>>>> guarantee of a zero being there.
>>>>
>>>> Hitting the oscillator makes it vibrate in all three axis, that's not going
>>>> to be suppressed regardless of which way you have it mounted.
>>>>
>>>> Bob
>>>> *********************
>>>> On Mar 28, 2010, at 12:42 AM, WarrenS wrote:
>>>>>
>>>>> Humm,
>>>>> Have to admit, I did not consider that as a possibility before.
>>>>> Maybe when I tap on it, its not microphonics after all that cause the freq
>>>>> to modulate, but the vibration of the inside stuff that is warming it up.
>>>>> For every action there is a reaction and for every nut there is a
>>>>> wing-it-nut.
>>>>>
>>>>> ws
>>>>> *****************
>>>>> Hi
>>>>>
>>>>> If that's the result you are getting, you are measuring something other
>>>>> than G sensitivity. Temperature effects possibly.
>>>>>
>>>>> Bob
>>>>>
>>>>> ****************
>>>>> On Mar 27, 2010, at 11:02 PM, WarrenS wrote:
>>>>>
>>>>>>
>>>>>> Just a friendly comment about the Zero G turn over point and Vibration
>>>>>>
>>>>>> Like Zero temp turn over, Special orientation of the OSC ONLY works good
>>>>>> over a VERY SMALL range, (maybe a 1/100 of G change)
>>>>>> It would not help vibration and has no effect on microphonics which are
>>>>>> likely a bigger problem anyway.
>>>>>> Try taping you Osc, It's freq will go crazy if monitoring it at high
>>>>>> resolutions and bandwidths
>>>>>>
>>>>>> ws
>>>>>>
>>>>>> ************************
>>>>>>
>>>>>> Hi
>>>>>>
>>>>>> The concrete basement floor is your friend.
>>>>>>
>>>>>> Stay as far away from the blower on the furnace as you can. If you have a
>>>>>> drop forge in the basement avoid it as well :)....
>>>>>>
>>>>>> You will indeed have a seismograph, but not a very useful one. There's
>>>>>> not a lot of G's at seismic frequencies unless you live in an active
>>>>>> earthquake region. The fundamentals of G's and displacement vs frequency
>>>>>> are in your favor in that respect.
>>>>>>
>>>>>> Bob
>>>>>>
>>>>>>
>>>>>> On Mar 27, 2010, at 12:42 PM, Stanley Reynolds wrote:
>>>>>>
>>>>>>> Is the source of the vibration important ? I'm thinking that any
>>>>>>> vibration that is not on the same axis as gravity. Walking across the
>>>>>>> lab vs a fan that is out of balance close by. Would a suspended mass
>>>>>>> mounting help with vibration isolation and damping with rubber pads and
>>>>>>> springs or would that just make a seismograph ?
>>>>>>>
>>>>>>> Stanley
>>>>>>>
>>>>>>>
>>>>>>> ----- Original Message ----
>>>>>>> From: Peter Vince <pvince at theiet.org>
>>>>>>> To: Discussion of precise time and frequency measurement <time-nuts at
>>>>>>> febo.com>
>>>>>>> Sent: Sat, March 27, 2010 10:51:07 AM
>>>>>>> Subject: Re: [time-nuts] Making a HP 10811 better
>>>>>>>
>>>>>>> Warren,
>>>>>>>
>>>>>>> If you turn over an oscillator, is the frequency change
>>>>>>> completely reversible (to your "under 1e-12 resolution") when it is
>>>>>>> restored? Thinking aloud, if an hour-glass is turned over twice, the
>>>>>>> final level will be the same, but the grains will be mixed. A quartz
>>>>>>> crystal, however, is solid, so hopefully nothing actually moves.
>>>>>>> Presumably the zero-G axis is with the axis of oscillation at 90
>>>>>>> degrees to gravity?
>>>>>>>
>>>>>>> Peter (the "other" one :-)
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>>> Another thing I use it for is to test high resolution Freq meters.
>>>>>>>> Using a calibrated wedge that I can then slide under one edge of the
>>>>>>>> zero-G
>>>>>>>> Osc box, I can
>>>>>>>> make small, variable, repeatable, freq changes of under 1e-12
>>>>>>>> resolution,
>>>>>>>> something pretty hard to do otherwise.
>>>>>>>> If I want to make BIG changes like 1e-10, I can rotate the box on any
>>>>>>>> of its
>>>>>>>> sides and still use the wedge,
>>>>>>>> and for a quick check of new equipment, I just turn the box over which
>>>>>>>> then
>>>>>>>> gives a couple of parts in 1e-9 freq change.
>>>>>>>> It makes a weird but simple and indispensable variable freq source that
>>>>>>>> is
>>>>>>>> useful for many things, such as checking the LOOP TC of a TBolt.
>>>>>
>>>>> _______________________________________________
>>>>> time-nuts mailing list -- time-nuts at febo.com
>>>>> To unsubscribe, go to
>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>> and follow the instructions there.
>>>>>
>>>>
>>>>
>>>>
>>>>
>>>>
>>>> _______________________________________________
>>>> time-nuts mailing list -- time-nuts at febo.com
>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>> and follow the instructions there.
>>>>
>>>
>>>
>>>
>>>
>>>
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
More information about the Time-nuts_lists.febo.com
mailing list