[time-nuts] OT: 5372A
Magnus Danielson
magnus at rubidium.dyndns.org
Fri Nov 19 23:15:32 UTC 2010
On 19/11/10 18:13, Pete Lancashire wrote:
> Thanks everyone. The question came up in what I am being charged for shipping
> From Texas to Oregon, $140 UPS Ground. Not complaining since I got the thing
> for $53. Hopefully will be FIP.
>
> -pete
>
> PS Thanks also on the battery issue. Next to get a real service manual.
I've done most of the things I need with what's on the net.
Replacement of the battery was trivial. Finding out that it was the root
cause was a bit less than obvious. The input sensitivity calibration
always goes flat-line, being a huge factor of annoying you, as you would
need to trim it up on each power-up. It's done with menues and no
service-mode required. Replacing the battery is a matter of finding a
3,6V Lithium battery and do some soldering. Then after power-up do the
input calibration and you are all set for the next 20 years or so. It is
worth the effort and not hard at all.
Cheers,
Magnus