[time-nuts] EIP545A 18GHz counter query
Ed Palmer
ed_palmer at sasktel.net
Tue Nov 27 18:40:06 UTC 2012
Hi Chris,
The first thing you should do is join the EIP_Microwave group at yahoo:
http://tech.groups.yahoo.com/group/EIP_Microwave
There's lots of info and help there for EIP counters.
Don't worry about the signature analyzer for now. It would normally be
used if the processor was dead. The fact that Band 1 works means that
the processor/memory is okay. If the processor / memory / display
related diagnostics like 02, 03, 04, 05 also pass, then signature
analysis is unlikely to tell you anything more.
The first thing you have to do is replace that missing chip. Bands 2
and 3 require the VCO and phase-lock circuitry to be working. As far as
I can see, U6 is not optional. Until the 200 MHz test passes, you can't
get any further. Be sure to check for obvious problems. I bought a
545A parts unit that wouldn't pass the 200 MHz test. It turned out that
some of the board-to-board cables were plugged into the wrong connectors!
I'm in the same position as you regarding testing at high frequencies.
You might be able to get a signal at the second or third harmonic of
your generator by cranking the level to the maximum and then using the
counter's Band 3 frequency limits feature to only look at that
frequency. My HP 8647A only goes to 1 GHz, but I can see harmonics up
to 3 GHz. I've also picked up a few cheap YIG oscillators that have
allowed me to test in the 1-10 GHz range. Watch out for the signal
levels. YIG oscillators have lots of output!
By the way, I'd be reluctant to purchase anything else from that
vendor. Chips don't evaporate overnight. Mind you, I guess chips have
fallen out of sockets. Any strange rattling noises when you shake it? :-)
Ed
On 11/27/2012 8:37 AM, Chris Wilson wrote:
>
> 27/11/2012 14:18
>
> I bought a 18GHz EIP545A counter which the vendor said was working
> fine the day before and when on overnight soak test, and also when
> last used some months agao. But when he checked it on the morning of
> my coming over to see it he found it had developed a fault... I bought
> it off him cheaply, as seen, hoping it might be fixable.
>
>
> Here's the basic tale:
>
> It was acquired displaying just dashes. I checked the PSU voltages and
> found the PSU section outputting well away from the manual figures. I
> corrected these and checked supplies for ripple and they were all
> good. I then decided to remove any board that was not essential to
> operation. It came with a GPIB option board, so I pulled it. Luckily
> this did some good and a working display appeared. There were three
> tantalum caps on this board, as a matter of interest I pulled a leg on
> each and tested them, all were OK. Refitting it killed the display
> back to "dashes" again, so I set it aside as having an "unknown
> fault", and continued. I can get it to accurately display up to
> EXACTLY 188.9999 MHz. Inputting 190.0000 MHz gives zeros. It has three
> frequency bands. Band one works perfectly right up and well beyond its
> 10Hz to 100MHz range. Band two often doesn't work at all, and just
> diplays zeros, it's a 50 ohm input, compared to the 1meg 20pF Band 1
> input. Sometimes by ramping the frequency up slowly from 100 MHz I can get a
> seemingly random reading. Band 3, the GHz range, doesn't work at
> all. Again, only zeros are displayed.
>
> The machine has quite a good range of self diagnosis tests. Tests are
> via the key pad. The first test checks the VCO and other stuff, and
> should display an accurate 200.0000 MHz. It displays well over, always
> in the range 253.5**** and the display isn't stable, it hovers over
> several KHz. There's a tree menu to see where this issue could lie.
> One limb suggests using a known good counter on the output of the VCO.
> I did this and the output is miles high in frequency, about double,
> and unstable. The tree menus goes on to suggest a phase lock circuitry
> fault.
>
> Further tests seem to depend in part upon using something called a
> Signature Analyser, which I have never even heard of They suggest an
> HP5004A, which is apparently pretty ancient. Is there a cheap way of
> acquiring something to do this sort of testing? I believe given a few
> pointers my scope, multi meter, sig gen and my other (working) Racal
> counter could take me further.
>
> It's a nice old thing, and I would quite like a means of counting into
> the higher frequencies it offers, but don't want to spend too much
> time or money on it. It works a damned sight more than when it first
> landed, which has kind of given me incentive to push a bit further,
> given I have a .pdf copy of the repair manual. Here's the page of the
> schematic I think is relevant, if it is a phase locked loop problem.
>
> This morning I realised there's a chip missing from a socket on the
> board in question. It's U6, a flip flop, part of the pre scaler I
> think? I assumed it was for some option, not fitted, but I am not so
> sure now. A Chinese Ebay seller is breaking one of these machines and
> he lists all the boards seperately, with decent photos. His board has
> this socket populated, and looking at the schematic, (linked at the
> bottom of this post), I think it's probably a vital component? I am
> sure the seller must have known about this, but who knows...
>
> I have ordered a new chip and will fit it whwen it arrives tomorrow.
>
> Now, assuming I ever get this thing up and running fully, is there a
> quick and dirty way of producing a test signal in the upper limits of
> its range to check it out, given my Marconi sig gen stops at 1040 MHz?
> Perhaps using a diode to give some harmonics?
>
> Diagram of the board I believe is faulty and missing the chip is at:
>
> http://www.gatesgarth.com/phaselockloop.jpg
>
> The text for this section, from the manual is at:
>
> http://www.gatesgarth.com/phaselocklooptext1.jpg
>
> and at
>
> http://www.gatesgarth.com/phaselocklooptext2.jpg
>
> Thanks for looking!
>
>
More information about the Time-nuts_lists.febo.com