[time-nuts] HP5065A C-field current is temperature sensitive
Bob Camp
kb8tq at n1k.org
Sun Aug 30 16:00:03 UTC 2015
Hi
Another counter argument against the “ratio mode” is that with zero current
through the wiper, odd things can happen. There is *always* something
else to consider in any real design.
Since they needed to calibrate the voltage swing, there aren’t a lot of options
with the technology they were using at the time. Today there are a lot more
choices of how to get this sort of job done.
Bob
> On Aug 30, 2015, at 8:05 AM, Adrian <rfnuts at arcor.de> wrote:
>
> Bob,
>
> agreed. But in the 5065A C-field circuit, the pot is wired as a variable
> resistor.
> The Bourns data sheets specifies +/- 50 ppm / deg C max.
> Together with the two series resistors, the effective TC of the pot
> drops to about 20 ppm.
> That is indeed just 2 times more than the effective temperature
> coefficient of the reference diode (1N938).
>
> Btw. the 1N938 is running at about 7.5 mA +/- 1 mA, not at 12 mA as
> mentioned earlier.
> Some 3.5 to 5.5 mA are going through R8, R6, R12.
> To get the correct diode current, this current has to be subtractrd from
> the total current through R7 (Q5, Q6 base currents neglected).
>
> Adrian
>
>
> Bob Camp schrieb:
>> Hi
>>
>> Most potentiometers have very good TC when used in a “ratio mode”.
>> You can quickly mess that up by putting resistance in series with
>> the pot. It’s a bit surprising how may times you see this inherent stability
>> messed up that way.
>>
>> Bob
>>
>>> On Aug 29, 2015, at 9:03 AM, Adrian <rfnuts at arcor.de> wrote:
>>>
>>> Keep in mind that potentiometer R6 is one of the key components of the
>>> C-field circuit.
>>> Together with R8 and R12, it defines the reference voltage to Q6A/B.
>>> The tempco of that wire-wound potentiometer can certainly not be neglected.
>>>
>>> I would suggest to check the C-field temperature dependency with an
>>> external precision current source feeding the C-coil directly.
>>>
>>> Adrian
>>>
>>> Poul-Henning Kamp schrieb:
>>>> Using the Fluke 732A as voltage reference improved HP5065A stability,
>>>> but not very much, so I looked closer at the constant-current
>>>> generator and the C-field pots "calibration":
>>>>
>>>> http://phk.freebsd.dk/hacks/HP5065A/20150828_c_pot/index.html
>>>>
>>>> ... and found out that the current-sensing resistors R10||R11 have
>>>> a very high temp-co.
>>>>
>>>> Getting closer...
>>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts at febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list