[time-nuts] HP5065A Step Recovery Diode, what part#?
ed breya
eb at telight.com
Tue Aug 28 00:03:53 UTC 2018
I forgot to mention:
If you've been diagnosing by swapping parts, be aware that changing the
A3 is the same as changing the SRD - they won't necessarily match
anymore, so would need adjustment.
Maybe your original SRD was just fine, but the whole set needed
adjustment. Who knows what may have been tweaked or mis-tweaked by
individuals working on it previously. If the original SRD is intact,
just check to see if it's still a diode (should look the same as a
regular Si diode, with maybe a little higher forward voltage drop). If
so, it's probably OK and can be returned to service.
Ed
More information about the Time-nuts_lists.febo.com
mailing list