OK, now I've got it.  I wasn't seeing the humps as being many slight variations around the 50ps points.  Just had my mental scaling all wrong.  In other words, the TICC's resolution is "blurred" around the nominal value rather than being aligned to a precise increment.

I wonder if it would help to round (or truncate) the TICC output to 10 ps, since the last digit is noise anyway.  Or go even further and round to the nearest 50 ps.

On Jan 8, 2019, 9:07 PM, at 9:07 PM, Tom Van Baak <tvb at leapsecond.com> wrote:
>> Wouldn't any counter that has 50 picosecond resolution look the same?
>
>
>Nope. If you had a 20 GHz timebase you'd have 50 ps resolution and all
>your readings would be exact multiples of 50 ps. It would be very clean
>and simple. Readings of 25 ps or 49 ps or any other non-multiple of 50
>ps would never occur. The histogram would be composed of thin vertical
>lines -- each exactly 50 ps apart.
>
>> Its data output is quantized by its resolution.  For example, the
>5334A 
>> with 2ns resolution will always show bins no less than 2ns apart.
>
>Right, the 5335A has a 500 MHz clock and so all readings are multiples
>of 2 ns. The counter cannot output a reading that ends in 1 ns or .123
>ns. The histogram would be composed of thin vertical lines -- each
>exactly 2 ns apart.
>
>Now it gets a bit more complicated with an interpolating counter, like
>a SR620 or hp5370. In this case there are fractions involved, but they
>are fixed fractions.
>
>The numerology [1] for SR620 quantization is something like 90 MHz,
>111.11 ns, 1/512, for a final result of 21.701... ps.
>The numerology [2] for the hp5370 is something like 200 MHz, 5 ns,
>257/256, 5.01953125 ns, 1/256, for a final result of 19.607... ps
>There is determinism or fine quantization in the readings because even
>though this ~20 ps value is not a nice round number, it is a fixed
>number.
>
>/// Consequently, for any of the 4 counters mentioned above, if you
>take a noise floor dataset and sort | uniq -c you will see a small
>number of large peaks. Like a picket fence. A picket fence inside a
>Gaussian envelope.
>
>What we see in the TICC is something quite different. It's not a
>problem, just very different. Why is that?
>
>> You've mentioned this before, and I'm having trouble getting my head 
>> around it.  I may have this all wrong, but isn't the quantization
>simply 
>> the resolution of the device?  Your histogram shows humps about 50 
>> picoseconds apart, but that's the resolution of the counter.
>
>Here's what's happening. Unlike the counters mentioned above there is
>no fixed cycle or fraction in the TICC. The ring counter is
>free-running and its rate can vary from unit to unit, from temperature
>to temperature, from reading to reading. That's why the all-important
>post-calibration step is performed for every reading. This results in
>readings that are not fixed fractions of 1 ns or 20 ps or 64 ps or 62.5
>ps or any other deterministic value. So each vertical "line" of the
>histogram is no longer a thin line, it is spread out quite a lot.
>
>/// If you take a TICC noise floor dataset and sort | uniq -c you will
>also see small number of large peaks, but the peaks are fat camel humps
>instead of thin picket fences. [3] It has a completely different look.
>Gaussian camel humps within a Gaussian envelope.
>
>/tvb
>
>[1] SR620 math:
>https://www.thinksrs.com/downloads/pdfs/manuals/SR620m.pdf
>[2] hp5370 math:
>http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1978-08.pdf
>[3] http://leapsecond.com/pages/ticc/
>
>
>----- Original Message ----- 
>From: "John Ackermann N8UR" <jra at febo.com>
>To: <time-nuts at lists.febo.com>
>Sent: Tuesday, January 08, 2019 1:07 PM
>Subject: Re: [time-nuts] Short term 10MHz source
>
>
>> You've mentioned this before, and I'm having trouble getting my head 
>> around it.  I may have this all wrong, but isn't the quantization
>simply 
>> the resolution of the device?  Your histogram shows humps about 50 
>> picoseconds apart, but that's the resolution of the counter.
>> 
>> Wouldn't any counter that has 50 picosecond resolution look the same?
>
>> Its data output is quantized by its resolution.  For example, the
>5334A 
>> with 2ns resolution will always show bins no less than 2ns apart.
>> 
>> What am I missing?
>> 
>> John
>> ----
>> 
>> On 1/8/19 3:34 PM, Tom Van Baak wrote:
>>> John,
>>> 
>>> Your hunch is correct. For most modern TIC devices, measurement
>noise is white. That is, if you do a self-test using a common DUT & REF
>you get a nice clean Gaussian plot. But the TICC is not like that. The
>TICC is based on a ring counter and so there is a *high* degree of
>quantization. This is not bad, per se, but it does impact how one
>should perform, or interpret, a noise floor test.
>>> 
>>> Take a look at: http://leapsecond.com/pages/ticc/
>>> 
>>> And in particular:
>http://leapsecond.com/pages/ticc/ticc-log5342-hist-1.gif
>>> 
>>> What this means is that a noise floor measurement made with the same
>chA and chB and REF could be quite wrong. This is not an indictment
>against the TICC. I have several, and use them all. But because of the
>strong quantization effects you can't just feed in a common DUT and REF
>and expect that to represent all possible real-life phase measurements.
>Those quantized "camel humps" are really quite extreme with the TICC.
>>> 