[time-nuts] A simple sampling DMTD
Bob kb8tq
kb8tq at n1k.org
Mon Feb 24 00:51:17 UTC 2020
Hi
Ok, but thats “high pass in the RF section”. You really do not have an audio high pass
filter the way you would in a more typical DMTD.
If it’s any comfort, I’m sitting here looking at a very different box. It also has “wobbles”
as you get into parts in 10^-16. That might change a bit if the draft coming through the
window was a bit less.
Bob
> On Feb 23, 2020, at 6:12 PM, Jan-Derk Bakker <jdbakker at gmail.com> wrote:
>
> Dear Bob,
>
> The capacitors are 47n NP0/C0G types (Kemet C0805C473K3GAC7800), picked for low tempco (and low DF and other non-ideal behavior). I've not spotted any hysteresis artefacts in these in previous designs, but I haven't measured their performance in this circuit.
>
> Forgot to mention in the previous message: the baluns are transformers (M/A-COM MABAES0060), so the only DC the ADC should see is its own input offset (plus offset current across the 25R input filter resistors). Full schematic is here ( http://www.lartmaker.nl/time-nuts/DMTD_rev0.99.pdf <http://www.lartmaker.nl/time-nuts/DMTD_rev0.99.pdf> ; needs cleanup, but all connections are there).
>
> JDB.
>
> On Sun, Feb 23, 2020 at 10:13 PM Bob kb8tq <kb8tq at n1k.org <mailto:kb8tq at n1k.org>> wrote:
> Hi
>
> What does the temperature coefficient of your “hardware HPF” filter caps look like?
> Are they a type that has significant hysteresis?
>
> Bob
>
> > On Feb 23, 2020, at 3:05 PM, Jan-Derk Bakker via time-nuts <time-nuts at lists.febo.com <mailto:time-nuts at lists.febo.com>> wrote:
> >
> > Dear Attila,
> >
> > Thanks for the heads up.
> >
> > I am currently using a HPF both in hardware (capacitive coupling into the
> > balun driving the ADC inputs) and in software before the ZCD. This should
> > counteract the first-order effects of this offset, although second-order
> > effects (converter nonlinearity et al) will of course still be an issue.
> > The plots you've quoted include (different kinds of) DC offset correction
> > for all but the "unfiltered" data; getting an efficient DC offset
> > correction working in real time on this 8-bit platform was indeed one of
> > the main challenges of the software-only approach.
> >
> > The FPGA daughterboard is currently in production at Eurocircuits; I hope
> > to have time to work on those the coming month. I'll also try to book some
> > time in our climate chamber. (I've had one of our GPSDO-designs running in
> > our general labs since before Christmas; surrounding it with bottles of
> > water works well enough to low pass filter temperature swings, but I still
> > see 6 degrees C swings overnight as out HVAC only runs during business
> > hours.)
> >
> > To be continued,
> >
> > JDB.
> >
> > On Sun, Feb 23, 2020 at 8:11 PM Attila Kinali via time-nuts <
> > time-nuts at lists.febo.com <mailto:time-nuts at lists.febo.com>> wrote:
> >
> >> Good evening!
> >>
> >> I'm going through some old stuff...
> >>
> >>
> >> On Wed, 27 Nov 2019 00:29:19 +0100
> >> Jan-Derk Bakker <jdbakker at gmail.com <mailto:jdbakker at gmail.com>> wrote:
> >>
> >>> This has yielded a combined "simple" signal
> >>> processing path of a differentiator, a double comb filter and the offset
> >>> estimator, which is getting very close in performance to the "ideal" band
> >>> pass filter (OADEV of 3.77e-13 at tau=1s versus 3.25e-13 at tau=1s for the
> >> BPF;
> >>> full plot:
> >>>
> >> http://www.lartmaker.nl/time-nuts/DMTD%20self-noise%20OADEV%20with%20PLL%20and%20various%20filters.pdf <http://www.lartmaker.nl/time-nuts/DMTD%20self-noise%20OADEV%20with%20PLL%20and%20various%20filters.pdf>
> >>> for this 600000-second recording:
> >>>
> >> http://www.lartmaker.nl/time-nuts/600ksec%20run%20with%20PLL,%2010811%20through%20splitter.png <http://www.lartmaker.nl/time-nuts/600ksec%20run%20with%20PLL,%2010811%20through%20splitter.png>
> >>> . OADEV past ~1000sec is severely compromised by the fact that the
> >>> measurement setup is in my home lab which sees temperature swings of up
> >> to
> >>> 20 degrees C and which does get bumped from time to time. Longer runs in
> >> a
> >>> more controlled setting forthcoming).
> >>
> >>
> >> I can offer an explanation for the large effect of the zero correction seen
> >> here. The LTC2140 is specified to have a +/-10µV/°C drift (at 1Vpp
> >> setting).
> >> Converted into phase error due to zero crossing shift, this turns into
> >> a phase shift of +/-1ps/°C @ 10MHz. Note, the shift is given as +/- and
> >> per channel, which means, it could very well be that the channels are
> >> not matched in their temperature characteristics and thus the total phase
> >> shift could be +/-2ps/°C ... though total shift being closer to 0.5ps/°C is
> >> more likely.
> >>
> >> Summa sumarum: DC offset correction is important if a zero crossing
> >> detector is used.
> >>
> >> Attila Kinali
> >>
> >> --
> >> <JaberWorky> The bad part of Zurich is where the degenerates
> >> throw DARK chocolate at you.
> >>
> >> _______________________________________________
> >> time-nuts mailing list -- time-nuts at lists.febo.com <mailto:time-nuts at lists.febo.com>
> >> To unsubscribe, go to
> >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com <http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com>
> >> and follow the instructions there.
> >>
> > _______________________________________________
> > time-nuts mailing list -- time-nuts at lists.febo.com <mailto:time-nuts at lists.febo.com>
> > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com <http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com>
> > and follow the instructions there.
>
More information about the Time-nuts_lists.febo.com
mailing list