[time-nuts] 5065 photo/lamp problem
Magnus Danielson
magnus at rubidium.se
Wed Jan 15 02:21:08 UTC 2020
Hi Paul,
Lacking the specific transistor 2N3553 I can for the moment not progress
with the rubidium lamp. It's very obvious that the oscillator is not
oscillate and hence the lamp is not giving light.
A friend might have those in his way to extensive component storage. It
has just not materialized so far. Life may interfere with digging around.
So, I moved a board over so I could run another 5065 and that way I can
do other experiments, with the aim to measure deeper on the maser, which
requires to make the 5065s better if possible. Turns out none of the
5065s is having very satisfactory phase-noise, and quick-and-dirty
attempt to fix that has improved the state, so it seems that it was for
sure the limiting factor for that 5065.
But that rubidium lamp shall be fixed and that 5065 become operational too.
Cheers,
Magnus
On 2020-01-15 00:40, paul swed wrote:
> magnus
> I was following the thread and it went quiet. Have you been able to make
> progress on the 5065? Always eager to learn, because one day my 5065 will
> fail.
> Regards
> Paul
>
> On Mon, Jan 6, 2020 at 5:04 PM <cdelect at juno.com> wrote:
>
>> Magnus,
>>
>> If you apply a +20V supply to the lamp connector you should see 125-150ma
>> unlit and it drops bit when it lights.
>>
>> Also measure from the lamp connector center pin to the outside conductor
>> with an Ohmmeter.
>>
>> You should have around 3.33K. The 1.33K resistor (in series with a 2.0K)
>> is a common problem. it can open up or rise in value.
>>
>> I usually replace this resistor in units I service even if it measures
>> good.
>>
>> I use a 1.3K 2% MOX resistor.
>>
>> Also check the approx. 12 Ohm resistor on the standoffs to see if it
>> opened up.
>>
>> If you pull the lamp assy. you can place a spec-A lead and see the RF.
>> Should be approx 88Mhz. (not critical)
>>
>> Once power is applied the lamp should start (cold) within a minute or
>> so, warm it starts fast.
>>
>> If you place a scope on A7 TP2 and with the lamp assy removed allow some
>> florescent lighting to fall into the interior you should see a large 50
>> or 60Hz signal from the light falling on the photocell.
>>
>> That is a quick check of the A7!
>>
>> Cheers,
>>
>> Corby
>>
>> Cheers,
>>
>> Corby
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at lists.febo.com
>> To unsubscribe, go to
>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>> and follow the instructions there.
>>
> _______________________________________________
> time-nuts mailing list -- time-nuts at lists.febo.com
> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list