[time-nuts] Re: Some assistance with Ball-Efratom FRK-L Rb Osc. Needed
rodger_adams at yahoo.com
rodger_adams at yahoo.com
Wed Sep 22 15:09:50 UTC 2021
In the last few years, I've obtained 3 FRK's and all 3 had the same issue.
The C-field voltage was so far off from where it needed to be that the servo
loop wouldn't lock up. In all 3 cases, I was able to get them working by
changing out resistors R17 and R20 in the C-field circuit on the power
supply board. Other resistors in the same area, R16, 18, and 19 might come
in to play too. In at least one of the units there was also a wire that
needed to be moved and possibly a jumper that had to be removed or added.
This is all very easy to do as it's on the top side of the power supply
board, and the resistors are mounted on terminals to make them easy to
change.
My guess is that the FRK's were installed in some larger system that used
the external EFC input and when the FRK's were pulled they were not
configured properly to be used standalone as I am using them.
I probably took some notes on exactly what I did to get them working. Let
me know if you need any details and I'll see if I can find (or remember)
exactly what I did.
Good luck,
Rodger
-----Original Message-----
From: ed breya <eb at telight.com>
Sent: Wednesday, September 22, 2021 2:15 AM
To: time-nuts at lists.febo.com
Subject: [time-nuts] Re: Some assistance with Ball-Efratom FRK-L Rb Osc.
Needed
If you look at the schematics of the C-field adjustment circuits, you'll
find where the tweaking is available for fine tuning the Rb system. I think
that's the only "frequency adjustment" spot to worry about. The other is the
tweak for the crystal oscillator to accommodate long term drift there, if
its tuning range hits a limit. This an internal, seldom needed adjustment,
so the C-field is likely the only one that would be useful externally, to
(very) fine tune the output frequency. I have some similar M-100 units,
which have access to the C-field pot through a hole in the back end heat
sink and case. An external adjustment port would be either a substitute for
the internal pot, or something to add to or subtract some current from the
C-field coil, that's mostly controlled by the internal pot.
So, look for an internal pot. If there is one, then the external stuff
likely just modifies what it does. If there isn't one at all, then the
external circuit may have to provide the proper current, so it may be a
little more complicated. It could also be that the C-field is set up to some
nominal value, with fixed resistors, without any internal pot, and an
external one just moves it around a bit.
If you just run it, and it locks, it should be good to go, and you can
figure out these fine tuning issues later.
Ed
_______________________________________________
time-nuts mailing list -- time-nuts at lists.febo.com -- To unsubscribe send an
email to time-nuts-leave at lists.febo.com To unsubscribe, go to and follow the
instructions there.
More information about the Time-nuts_lists.febo.com
mailing list