[time-nuts] Re: Testing frequency pulling on a DYI counter

Erik Kaashoek erik at kaashoek.com
Sat Aug 6 07:15:06 UTC 2022


Magnus,

Many thanks for the reference to your article on linear regression. I 
will need a lot of time to study this as my math skills have become very 
rusty after not having been used for 40 years.
Not sure what you meant with "cancelation"

What I did is much much more simple.
The Excel simulation is an easy way to understand the calculations [1]
Let me know if you can not open an Excel file.

The input data is in red cells
The important output is in blue cells (phase, frequency)
The formula are in green cells
The "time noise" input was to check if dithering of the interpolation 
points over time could improve the outcome, which it did not.

The internal ref frequency is derived from a 10MHz reference using a PLL

In the real code all sums are calculated in 32 or 64 bit integers. This 
limits the maximum gate time to somewhere above 10 seconds.
The simulation uses 0.1 ms interval between the interpolation points, 
the actual implementation uses 0.05 ms
At the gate moment the final divide into a 64 bit double is done, if 
sufficient interpolation points, otherwise the direct calculation is used.

As you can see in the Regression error graph the final gate moment has a 
big influence on the error which is a pity as the pattern in the error 
suggests there is something that can be "interpolated" or eliminated
And the impact of the final gate moment increases when close to an 
integer multiply/divide relation to the reference.

Hope, once I understand your algorithm, this can be improved.

[1] http://athome.kaashoek.com/time-nuts/Regression.xlsx

On 5-8-2022 22:28, Magnus Danielson wrote:
> Erik,
>
> Which algorithm of linear regression do you use? Describe the 
> cancellation details.
>
> I just want to be sure we talk the same language here.
>
> Have you seen this? https://arxiv.org/abs/1604.01004
>
> Cheers,
> Magnus
>
> On 8/4/22 20:04, Erik Kaashoek wrote:
>> Bob, Magnus,
>>
>> Using a second counter (my famous Picotest U6200A) locked to the 
>> reference output of the DIY counter and measuring the output of the 
>> signal generator and also set to gate of 10 s it is confirmed that 
>> the frequency pulling (if any) is below 1E-11 (not more digits on the 
>> display of the U6200A)
>> Generator is set to 10.000,000,000,2 MHz and is measured as such by 
>> the U6200A
>> As there seems to be no frequency pulling I went back to the 
>> simulation of the linear regression algorithm and discovered that 
>> when there is a integer  divide/multiply relation between the 
>> internal reference and the measured frequency the regression looses 
>> some accuracy.
>> For sure if the reference is close to an integer multiple of the 
>> measured frequency (10 Mhz measured -> 200 MHz reference) the 
>> regression collapses completely in accuracy. I hoped that by creating 
>> a fractional relation this collapse would not happen at 10 MHz but is 
>> still there, although much smaller. For this test I'm using a "div 3 
>> times 64 e.g. 213.333,333,333,333... MHz" internal reference 
>> frequency derived from the external 10MHz reference. Ton van Baak 
>> warned me against using fractional relations in a counter but 
>> otherwise it is impossible to measure a 10 MHz input signal with any 
>> accuracy without a HW time to digital as the interpolation no longer 
>> works. I can switch dynamically to 200 MHz or 245 MHz reference and 
>> these produce much much worse results.
>> I realize this test only measures if the TCXO used as reference in 
>> the DIY counter does not show frequency pulling but it does not show 
>> if the PLL used to convert the 10MHz to 213.333333333... MHz for the 
>> internal counters shows any frequency pulling.
>> NOt
>> Erik.