[time-nuts] Problems and solution when using a Rigol DDS signal generator for counter testing
Erik Kaashoek
erik at kaashoek.com
Fri Aug 19 08:32:22 UTC 2022
During the tests of a frequency counter using a Rigol DG812 signal
generator locked to an external 10MHz clock it was observed that the
output frequency of the DG812 was not stable
This can clearly be seen in [1], blue trace, where the counter was
warming up and the frequency of the frequency variations changed over time.
To test if the external locking was related to the instability the
external lock was disabled which lead to huge instability. [2] Pink trace
As can be seen in the pink trace the huge variations disappeared when
the coax cable with the 10MHz reference clock was disconnected (at 48
seconds) and appeared again when the cable was reconnected to the clock
input (at 86 seconds), regardless of the Rigol running the whole time on
its internal clock.
This seemed to hint at some interference between the external clock
signal and the internal frequency reference happening when running both
at internal or external clock.
To test this an external clock signal was applied at the frequency (+/-
0.01 Hz ) of the internal clock.
The instability of the output from the Rigol locked to an external clock
close to is internal frequency disappeared completely ([1], red trace)
and the remaining frequency variations where within the measurement
uncertainty of the used frequency counter.
[1] http://athome.kaashoek.com/time-nuts/Rigol%20locking%202.png
[2] http://athome.kaashoek.com/time-nuts/Rigol%20pulling.png