Luciano:
I would like to try the amplifier for my 10MHz output.
Where can I find details for the toroid transformer,
how many windings, and what gauge wire?
Thanks,
Neal
N6YFM
On Sun, Apr 23, 2023 at 6:32â¯AM Luciano via time-nuts <
time-nuts at lists.febo.com> wrote:
>
> Hi Matt,
>
> I've built a buffer amplifier just for the use you want.
>
> The design comes from Bruce and was optimized and engineered by me and
> a group of friends.
>
> You can see the solution at the addresses I enclose.
>
>
> http://www.timeok.it/wp-content/uploads/2015/08/Frequency-Distribution-Design-Basic-Module-v-4.1.pdf
>
>
> http://www.timeok.it/wp-content/uploads/2015/08/hp5065A-option-H10-v-2.00-2.pdf
>
> Compared to the one you propose, it has several improvements, such as a
> higher input impedance and a much higher reverse isolation that exceeds
> 100dB.
>
> I hope this info can be useful to you,
>
> Luciano
>
> Luciano P. S. Paramithiotti
> timeok at timeok.it
> www.timeok.it
>
> Da "Matt Huszagh via time-nuts" time-nuts at lists.febo.com
> A "Discussion of precise time and frequency measurement"
> time-nuts at lists.febo.com
> Cc "Matt Huszagh" huszaghmatt at gmail.com
> Data Sat, 22 Apr 2023 18:21:07 -0700
> Oggetto [time-nuts] Bruce's common emitter transformer feedback
> amplifier: oscillations with capacitive load
> Hi,
>
> I built Bruce Griffith's common emitter transformer feedback amplifier
> (http://www.ko4bb.com/~bruce/OCXOBufferAmplifiers.html). My build is
> mostly identical to the first schematic in the previous link. But, I
> added a second stage for increased reverse isolation. I placed 65R
> across the first stage output, which gave me about the right gain in
> simulations. I also added a second 2-stage buffer in parallel. I
> custom-wound the transformer on a Fair-Rite 67 toroid. I'm planning to
> use this to get buffered/isolated front and rear 10 MHz outputs from the
> internal 10811A in my HP 5065A.
>
> Overall, the performance seems excellent. I've attached a residual phase
> noise measurement, showing the combined 2 cascaded stages. These results
> seem only a little worse than what John Miles measured for his single
> stage (http://www.ke5fx.com/norton.htm).
>
> I measured the reverse isolation to be 76 dB and 77 dB and the isolation
> between outputs to be 49 dB or 59 dB, depending on which output I inject
> the signal into. This is all at 10 MHz of course.
>
> However, I am experiencing one issue, which is that if I add a
> capacitive load to the output (such as from a coaxial cable), I get
> large oscillations (10+Vrms) at the output. The oscillations are at
> about 20MHz for a 1m long RG400 cable (94pF/m). These are not present
> when the output is unloaded, nor when it's loaded by the typical
> 50ohms. Since this only appeared when I added a capacitive load to the
> output, I figured it was due to the parallel LC between the output
> winding and coaxial capacitance. My output impedance came out a bit
> lower than I'd wanted (35ohms), so I tried inserting 10ohms in series
> between the output winding and coaxial connector. This cut the
> oscillations nearly in half (now about 6.5Vrms), but didn't kill
> it. I've also tried loading down the output with a 1k parallel resistor,
> but though this helps too, the oscillations are still pretty
> large. Going lower helps further, but I don't want to load down the
> output too much.
>
> Any thoughts? Has anyone else seen this behavior? I can post more
> details and images.
>
> Thanks,
> Matt
> _______________________________________________
> time-nuts mailing list -- time-nuts at lists.febo.com
> To unsubscribe send an email to time-nuts-leave at lists.febo.com
>
More information about the Time-nuts_lists.febo.com
mailing list