[time-nuts] Re: Faulty HP 10811D
Denis Dowling
dpd at opsol.com.au
Wed Mar 29 10:16:47 UTC 2023
Thanks everyone for all of the advice. As predicted the fault was an
open circuit thermal fuse. I have bypassed this (temporarily) and I can
now trim the oscillator to be 10MHz to the precision of my PM 6672
frequency counter. I am plotting both 10MHz waveforms on a scope and
there is minimal slip. I will order a new fuse and then reassemble and
then attempt a more accurate calibration when everything has been
allowed to burn in for a while.
On a related note, how good is the Hewlett Packet March 1981 Journal
article on the HP 10811D?
http://hparchive.com/Journals/HPJ-1981-03.pdf
There is so much information in this article including how they used a
flex PCB to make the oscillator easier to assemble. They even discuss
the issues they had with stray capacitance due to the thinner PCB and
tearing on sharp edges. Unfortunately today I doubt any manufacturer
would be as open about their design process.
Regards,
Denis
On 28/03/2023 6:44 pm, Adrian Godwin via time-nuts wrote:
> I had a 10811 that didn't warm up. I'd never taken one apart before but,
> although a bunch of parts had failed, they were all common (if ancient)
> op-amps and the like. I didn't have much trouble fixing it once I'd got
> started. I'd encourage you to go ahead.
>
> On Tue, Mar 28, 2023 at 7:28â¯AM Askild via time-nuts <
> time-nuts at lists.febo.com> wrote:
>
>> Hi Dennis,
>>
>> The HP 10811 are serviceable. Just download the service manual and study
>> it.
>> I would expect the oven to be faulty with this much error.
>> Most likely the thermal fuse has gone open circuit.
>> The fuse can go open circuit without there actually being any fault with
>> the oven, but there might also be a fault in the oven circuit.
>> The fuse is on the first PCB you get to after opening up the 10811. All
>> described in the manual.
>> If its the thermal fuse, replace with correct thermal fuse. When replacing
>> make sure to not over heat it while soldering.
>> You can of course bridge the fuse with a wire for testing, but then you
>> need to monitor it closely, to be sure it does not over heat.
>>
>> I was lucky to find a HP 8663A in a electronic recycling bin. It was
>> missing PCB's in the PSU and the 10811 was totally burnt to a crisp.
>> Guess the fuse had been bridged. So I needed to get a replacement off ebay.
>>
>> Regards,
>> Askild
>>
>>
>>
>> On Tue, Mar 28, 2023 at 7:49â¯AM Denis Dowling via time-nuts <
>> time-nuts at lists.febo.com> wrote:
>>
>>> Hi All,
>>>
>>> This is my first time on the list so apologies if I am going over an old
>>> topic.
>>>
>>> I am repairing a HP 8662A Synthesised Signal Generator that I purchased
>>> from a government auction. The unit was purchased without any
>>> information about its working state and I have fixing various issues
>>> with it. Firstly there was a digital logic problem with the keyboard
>>> card that was causing the generator to periodically lock up. When this
>>> was fixed the unit came to life but the output signal level was down 5dB
>>> and this was traced to a bad step attenuator. I attempted a fix but
>>> ended up needing to replace this to get the output steps correct.
>>>
>>> I have now turned my attention to calibrating the reference timebase and
>>> unfortunately this seems to be well out of spec.
>>>
>>> The timebase is a HP 10811D 10MHz OCXO. The 10MHz output of the
>>> oscillator seems to be ~40Hz low. I have attempted to adjust the unit
>>> via the multi-turn capacitor adjustment but this seems to be at the end
>>> stop and the frequency is still low. So my questions are:
>>>
>>> 1) I assume this module is faulty as it is so far out of spec? The
>>> signal generator has been powered off for multiple months but even after
>>> having the unit powered on for 24 hours the frequency has not changed
>>> significantly.
>>>
>>> 2) How repairable are these modules? I ended up blowing $200 on the
>>> replacement step attenuator so will need to save up for any new parts. I
>>> can always use an external 10MHz reference but it is always nice to have
>>> equipment complete.
>>>
>>> 3) From searching the list it seems that the oven could be what is
>>> faulty. Would a cold over account for this much error?I have confirmed
>>> that the edge connector is supplying 25V on the over supply connections
>>> but I don't think the module is warming up.
>>>
>>> 4) Is it possible the OCXO has just worn out and has aged beyond the
>>> point it can be adjusted back into spec?
>>>
>>>
>>> Regards,
>>> Denis
>>>
>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts at lists.febo.com
>>> To unsubscribe send an email to time-nuts-leave at lists.febo.com
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at lists.febo.com
>> To unsubscribe send an email to time-nuts-leave at lists.febo.com
> _______________________________________________
> time-nuts mailing list -- time-nuts at lists.febo.com
> To unsubscribe send an email to time-nuts-leave at lists.febo.com
More information about the Time-nuts_lists.febo.com
mailing list