[volt-nuts] Quality Multimeters: HP 3457A
Dr. Frank Stellmach
drfrank.stellmach at freenet.de
Sun Aug 2 19:50:40 UTC 2009
Hello Roy,
take into account the 34401A also, as the spec (stability) are very
similar to the 3457A.
The 3457A has some extended DCI ranges, compared to the 34401, but also
no complete ACAL feature, as the 3458A.
I have in mind, that the 3457A may also be used in 7 1/2 digit mode, but
could not find this in the manual.
The 34401A can deliver more digits than 6 1/2 via GPIB
Linearity is 2ppm, but the 3457A is not specified at all in this regard.
Seems to me, that the 3457A is an elderly design, discretely built, and
has not been so common.
The 34401A design is derived from the 3458A, more compact and ruggedized.
Anyhow, better go for the 3458A.
ACAL, its stability and precision, the complete set of current ranges,
the different possibilities to make high precision AC measurements, and
the digitizing fetures are all worth the money.
I was lucky to pick a 9 year old unit here in Germany for about 3000€
recently.
I built a LTZ1000A based, dual reference, as a replacement of Fluke 732A
/ 732B or 7001.
Short term stability (1h) compared with the 3458A is around 2e-7.
Then I added a one decade precision divider for linearity testing (like
a 720A), and a 10:1, 100:1 reference divider, like the 752A.
Mentioned commercial devices are also rare here, but if you find them,
they are too expensive in most cases here in EU. I think, the US market
would be a better place to acquire them, despite tax & toll.
The analogous device to the time-nuts cesium clock would be a Josephson
array plus HF-devices plus liquid helium. It's a pity that "volt" cannot
be transferred wireless also..
Frank Stellmach
More information about the volt-nuts
mailing list