[volt-nuts] HP 3458A SCAL
Greg Burnett
gbusg at comcast.net
Wed Jun 23 08:38:58 UTC 2010
...In case anyone wants to try Swerlein's Algorithm, here's a link to all
the downloads:
http://www.home.agilent.com/agilent/editorial.jspx?cc=US&lc=eng&ckey=235459&id=235459
Greg
----- Original Message -----
From: "Greg Burnett" <gbusg at comcast.net>
To: "Discussion of precise voltage measurement" <volt-nuts at febo.com>
Sent: Tuesday, June 22, 2010 4:50 PM
Subject: Re: [volt-nuts] HP 3458A SCAL
Dick,
I think the mode you're referring to is its ACV SYNC mode (Synchronously
Sub-sampled Mode) which can be used to measure repetitive AC waveforms from
1Hz to 10MHz.
Here are some excellent articles:
HP Journal April 1989 (Several excellent 3458A articles)
http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf
Ron Swerleins' 10ppm Accurate Digital AC Measurement Algorithm
http://www.home.agilent.com/upload/cmc_upload/All/Swerleins_Algorithm.pdf
Note: Ron's "God's AC" algorithm uses the 3458A's DCV mode to sample AC
waveforms, and is good for AC waveform frequencies at or below 1kHz.
Best,
Greg
----- Original Message -----
From: "Dick Moore" <richiem at hughes.net>
To: <volt-nuts at febo.com>
Sent: Tuesday, June 22, 2010 2:02 PM
Subject: Re: [volt-nuts] HP 3458A SCAL
One of the joys of the 3458 is that at 100kHz and below, you can tell the
box to bypass the AC rectifier circuit and directly sample the AC waveform
with the ADC and then do all the processing, which yields quite good --
well, very good -- accuracy. Above 100kHz, the system must be used like the
majority of DMMs and do an absolute value conversion before ADC and
computing the TRMS. So I think your system will provide good results either
way at, say 500kHz and below, and the 2MHz and 8MHz flatness will have
essentially no affect on either way of processing the signal at lower
frequencies.
Dick Moore