[volt-nuts] Keithley 2001 Multimeter Fault
John Lofgren
jlofgren at lsr.com
Fri Mar 5 16:09:51 UTC 2010
Yes, check capacitors. I has something similar happen to the -5V regulator on HP 55300A GPSDO. The symptoms were very similar to your description, but with the addition of a discolored heatsink. It turned out to be a shorted ceramic cap on the regulator output.
Good luck and keep us posted.
- John
-----Original Message-----
From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com] On Behalf Of Laurence Motteram
Sent: Thursday, March 04, 2010 7:04 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Keithley 2001 Multimeter Fault
Don't be misled by the working display, as it has its own
microprocessor. The digital PCB has two LM2940CT devices, but it sounds
like the one in question supplies the digital +5V. Naturally,
everything connects to this, but you might like to start by checking
capacitors. If you are lucky, you could find a shorted tantalum.
Regards,
Laurence Motteram
-----Original Message-----
From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com] On
Behalf Of Alan Scrimgeour
Sent: Friday, 5 March 2010 11:32 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Keithley 2001 Multimeter Fault
The little functionality the meter had when I got it enabled me to start
a
self test, but now I can't even do that as functionality has dropped
away
rapidly. The display is fine and presumably the processor driving it,
but it
no longer responds to the keyboard.
A smell of overheating plastic after a few minutes lead me to an
LM2940CT -
5 volt regulator. It's located in the fan assisted cooling path but
getting
far too hot even allowing for that. It's output voltage is well below
5V.
It's input voltage appeared to be reasonable given it's being
overloaded, so
I conclude that either the LM2940CT is faulty or more likely, something
is
overloading its output. A circuit diagram would show me what connects to
this output so I can start searching for the culprit.
Alan
----- Original Message -----
From: "Laurence Motteram" <LMotteram at scientific-devices.com.au>
To: "Discussion of precise voltage measurement" <volt-nuts at febo.com>
Sent: Thursday, March 04, 2010 9:34 PM
Subject: Re: [volt-nuts] Keithley 2001 Multimeter Fault
> IP issues aside, what seems to be the trouble with the 2001?
>
>
> Regards,
>
> Laurence Motteram
>
> -----Original Message-----
> From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com]
On
> Behalf Of Alan Scrimgeour
> Sent: Friday, 5 March 2010 5:47 AM
> To: Discussion of precise voltage measurement
> Subject: [volt-nuts] Keithley 2001 Multimeter Fault
>
> I'm trying to repair a 7 1/2 digit Keithley Multimeter but can't get
the
> circuit diagram. Keithley say they don't release schematics any more
due
> to IP issues, whatever they are?
>
> Can anyone help?
>
> Thanks,
>
> Alan
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
_______________________________________________
volt-nuts mailing list -- volt-nuts at febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts