Charlie wrote:
>I have some laboratory grade voltage standards here in my home
>lab. I have a Fluke 5440B direct voltage calibrator which is
>cross-checked and compared against a pair of Fluke 732A DC Reference
>Standards that run on a UPS and are kept hot all the time. My
>ultimate transfer uncertainty at 10 volts should be less than +/-
>6.0 ppm per year and less than +/- 1.5ppm per 90 days over a
>temperature range of 18 degrees C to 28 degrees C based on the Fluke
>732A specifications.
>
>The only problem is that I will not ship these units because that is
>bad for calibration, stability, and uncertainty. However, something
>could be arranged if you want to ship the unit to be calibrated to me.
My understanding of Bob's original proposal was that he wants to
build a stable reference of uncertain voltage, and send that to
someone (or several someones) who will render an opinion on its
actual voltage and send it back. A well-calibrated 5440B and a null
voltmeter (Fluke 845A/AB/AR, HP419A) would work, as would a 732A or
equivalent and a precision voltage divider (Fluke 720A or equivalent).
Bob did mention having the recipient power the traveling reference up
-- if possible, it should be shipped "hot" under battery power so
there is no power cycle/retrace to worry about.
Best regards,
Charles