Bob wrote:
>Power transformer / rectifier / capacitor => ~35 Volts DC.
>LM7815A => 15 Volts (regulated).
With 20 volts of overhead, why not use two stages of pre-regulation
(say, LM7824/LM7815) to improve the line regulation even
further? You might also use separate LM7815s for the reference and
heater supplies, to minimize the reference supply changes due to
varying heater current and the associated load regulation/wiring drop.
>The two largest errors (other than 20 ppm stability of the lm199ah)
>are both under 1 ppm. They could be reduce further but is it worth
>it given the stability of the lm199a?
It is good design practice to reduce all sources of error to the
degree that they can can be reduced for little cost. In this
context, I consider two additional three-terminal regulators "little
cost." (If two additional regulators are just too much to bear, the
heater could be run from the 24 V supply.)
BTW, here is a standard similar to what you are
proposing: http://gellerlabs.com/LNVR%20Series.htm It uses an LM399
and the scaling amplifier of an AD587 to produce 10 V (I have
previously criticized this design for using the 587 instead of a
discrete op-amp that would give better performance, but the 587 used
this way appears to meet your design goals). Geller also offers a
$40 bare board based on the 587 alone, for which he claims short-term
transfer accuracy of +/- 5
ppm: http://gellerlabs.com/SVR%20Series.htm Here is another one
(+/- 25 ppm at 5
V): http://www.voltagestandard.com/New_Products.html I believe
that both vendors will recalibrate them for a nominal fee.
Best regards,
Charles