Charles,

Just a suggestion.  If you post where you are located, perhaps a list member with
better capability might announce himself and offer to help.  Also, depending upon
where you are located, less expensive alternatives might be available in the form
of local calibration facilities.

Failing all that, and depending upon your actual needs, perhaps obtaining a used
Fluke 515 (inexpensive calibrator) would serve you better ?  Then a one time
expense for calibration of the 515 would serve for all your instrument needs.

Of course, if you are not careful, you can sucked into the game and spend a lot
of money on other more expensive calibration equipment to put you mind at ease.
After all, Timenuts and Voltnuts are obsessive people.

Bill....WB6BNQ
(Located in San Diego)


Charles wrote:

> First post :)
>
> Having recently repaired a nonworking $5 hamfest special HP
> 34702A/34740A (4-1/2 digit DVM/display unit), I am interested in
> calibrating it. Without paying many times my purchase price to a
> lab!
>
> I obtained some sample MAX6325EPA (2.500V +/- 0.02%) which gets me
> "pretty close" for the 10V range. I also have a handy 596.5 ohm,
> 0.01% resistor for the Ohms ranges.
>
> The basic DC accuracy spec for the HP is something like +/- 0.03%
> plus 0.01% of range, so I need a better standard than that. Also I
> don't have a good way of generating very stable 10 KHz AC signals
> to cal those ranges...
>
> Much of the discussion on this list is 7.5 or 8.5 digit meters, so
> a 4.5 should be easy for you "Nuts". Any tips?
> thanks
> -Charles
>