[volt-nuts] Solartron 7081, other meters, and some voltage standards
gbusg
gbusg at comcast.net
Fri Sep 2 23:51:43 UTC 2011
Hi Joe,
To display one more digit of resolution on your HP 3457A, do this:
1. Press "DIGITS DISP"
2. Press "6"
3. Press "ENT"
For a more quiet reading, do this:
1. Press "NPLC"
2. Press "100"
3. Press "ENT"
After several hours warm-up, be sure to run "ACAL ALL".
I'm guessing your HP 3457A might typically be within 10ppm at 10Vdc.
Possibly within 5ppm but not likely better than that due to drift and
stability issues.
For example, my 3457A has drifted 5ppm since last "CAL 10" (5 years ago).
Of course in your case there's the uncertainty of not knowing when your
instruments were last cal'd by previous owners.
It sounds like you're a prime candidate to participate in a VDC round-robin,
as the other fellows are now discussing and proposing?
Cheers!
Greg
----- Original Message -----
From: "J. L. Trantham" <jltran at att.net>
To: "Discussion of precise voltage measurement" <volt-nuts at febo.com>
Sent: Friday, September 02, 2011 5:02 PM
Subject: [volt-nuts] Solartron 7081, other meters, and some voltage
standards
I have been testing my 'new' 7081 and suspect I have arrived at the 'usual'
cross roads of what to believe.
I have two Fluke 731B's that I got from 'the site'. I replaced the battery
packs in months past and have left both on for the past 3 or 4 days along
with the 7081 and other meters, an HP 3478A and an HP 3457A, as well as a
Fluke 335A voltage standard set to 10.000000 volts, all from 'the site'.
Regarding the 731B's, one is a 'white' face and the other is a 'black' face
and both are set to 10V.
I left the 7081 connected to the 'black' face 731B, set to 7 X 9 (3.2s) in
'TRACK', and walked in to the shop to find it reading 10.000000. When
connected to the 'white' face 731B, it reads 9.999898 and, when connected to
the 335A, it reads 10.000392.
The 3478A reads 10.0001, 10.0000, and 10.0005, respectively. The 3457A
reads 10.0001, 10.0000, and 10.0005, respectively.
Soooooooo, what, if any, conclusions are reachable? I am tempted to believe
that the 'black' face 731B and the 7081 are correct. What are the odds?
Thanks in advance.
Joe