I measured a 731 with a scope to see the noise. It was around 25 uVpp. I think that is the most easy way. You see how much disconnected and how much it s connected. I úse a tektronix 547 with E-plugin for this. ( a lot cheaper then a 7,5 digit or more multimeter ;-)
Fred PA4TIM
Op 27 sep. 2011 om 05:04 heeft "gbusg" <gbusg at comcast.net> het volgende geschreven:
> Hi Joe,
>
> My calculations agree with yours, as far as Fluke specs for your two
> voltages are concerned.
>
> Fluke spec'd Ripple and Noise (all frequencies) as <20 uV rms on the 10V
> range. So that would be another 2ppm rms adder for 10Vdc setting, 20ppm rms
> adder for 1Vdc setting and 200ppm rms adder for 100mVdc setting. But I'm not
> immediately sure how that rms noise spec might translate to observed
> racking-around on your 3478A (depending on integration time setting on your
> 3478A, etc.).
>
> Fluke's basic 10 uV adder (for 10Vdc range) translates to 1ppm adder at
> 10Vdc setting, 10ppm adder at 1Vdc setting and 100ppm adder at 100mVdc
> setting. The way I look at it, that 10 uV adder accounts for some short-term
> instability (which might also show up as some racking around on your 3478A)
> plus DC zero offset (as observed on your 3478A).
>
> Anyway, obviously the 335D "likes" full (or near full) scale the best.
>
> You could test your 3478A short-term noise (racking around of readings) by
> using it to read a source that's known to be very quiet. Or at least you
> could build a 10 to 1 or 100 to 1 resistive divider between 335D and your
> 3478A to see if that reduces the observed noise.
>
> Best,
> Greg
>
>
> ----- Original Message -----
> From: "J. L. Trantham" <jltran at att.net>
> To: "'Discussion of precise voltage measurement'" <volt-nuts at febo.com>
> Sent: Monday, September 26, 2011 8:39 PM
> Subject: Re: [volt-nuts] Fluke 335A
>
>
> Greg,
>
> If I understand the spec's correctly, the accuracy at 10.000000 Volts is +/-
> 30 uV (+/- (0.002% of setting + 10 uV)). At 0.123456 Volts, it would be +/-
> 12.5 uV. The stability spec is +/-(0.001% of setting + 10 uV) per month or
> 11.23456 uV. I think I am within those specs.
>
> However, I did not see a 'noise' spec.
>
> The question is how stable should the reading be when watching it on a DMM?
> Mine fluctuates about 3 uV over about 2 seconds (on an HP 3478A at nominal
> settings). Is that normal? I am hoping someone with a 335A or D will be
> able to look at theirs and tell me if what I am seeing is reasonable or do I
> have more work to do to obtain better stability/lower noise.
>
> It ultimately boils down to what to trust, the meter or the standard. You
> know the old saying, 'a man with one volt meter......'.
>
> Thanks,
>
> Joe
>
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.