[volt-nuts] Fwd: Solartron 7150plus - Chasing AC gremlins

Wim de Jong wim.de.jong.59 at solcon.nl
Fri Dec 7 12:19:47 UTC 2012


Dave,

Thats the question. It depends on the quality of the cells where the AC 
calibration info is stored.
I'm not aware of a checksum function on these EAROM's.

If it is the AC part, I've some spare parts if necessary... Just let me 
know.

Kind regards, Wim.

Wim de Jong
wim.de.jong.59 at solcon.nl

Op 7-12-2012 10:25, David C. Partridge schreef:
> Wim
>
> I don't think that is likely, as Phil clearly says that all the DC ranges are correct,  it is just the AC ranges that are a problem.  If the EAROMs were bad, you'd lose all the calibration constants.
>
> Dave
> -----Original Message-----
> From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com] On Behalf Of Wim de Jong
> Sent: 07 December 2012 08:04
> To: volt-nuts at febo.com
> Subject: [volt-nuts] Fwd: Solartron 7150plus - Chasing AC gremlins
>
> Phil,
> I suspect it to be the EAROM(s) IC 105/6 of this meter. Losing its calibration constants.
> Regards, Wim.
>
> Wim de Jong
> wim.de.jong.59 at solcon.nl
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>



More information about the volt-nuts mailing list