[volt-nuts] Solartron 7150plus - Chasing AC gremlins
m k
m1k3k1 at hotmail.com
Fri Dec 14 18:51:41 UTC 2012
Hi,
with several rms-dc converters inside dvm's leads me to find that there is a tradeoff between low level linearity and setting to the lowest levels. check the output at 1% of scale vs the zero offset.
M K
> From: david.partridge at perdrix.co.uk
> To: volt-nuts at febo.com
> Date: Fri, 14 Dec 2012 06:09:32 +0000
> Subject: Re: [volt-nuts] Solartron 7150plus - Chasing AC gremlins
>
> Adjustment of the offset trim should be trivial - short inputs on an AC range, and adjust offset trim to give lowest possible uV DC reading at TP3.
>
> I just checked the Service Manual and that's exactly what page 3.6 tells you to do (Test 7) - it says to use 20V AC range.
>
> Where did you find LH0062CH at a civilized price!???
>
> Dave
> -----Original Message-----
> From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com] On Behalf Of Philip Pemberton
> Sent: 14 December 2012 01:44
> To: Discussion of precise voltage measurement
> Subject: Re: [volt-nuts] Solartron 7150plus - Chasing AC gremlins
>
> On 13/12/12 17:32, Rob Klein wrote:
> > Phil,
> >
> > However contradictory it may sound, that is actually a good thing.
> > Continuous
> > failures are always easier to chase down than your gremlins.
> >
> > Tracing front to back should bring you to Mount Doom relatively
> > quickly ;-)
>
> Indeed!
>
> Well, here's the sum total of this evening's testing:
>
> * Disconnecting LK2 is not a good idea. It appears to connect the A/D converter up to the input loop. Disconnecting it kills all the measurement ranges (they'll read a constant low voltage).
> I thought it was for disconnecting the drift-correction circuitry from the input loop... c'est la vie.
>
> * Lifting one leg of R82 (the 75R input resistor for the RMS-DC
> converter) and injecting 2V pk-pk 400Hz from a sine generator (Zo=50R) results in a stable reading of ~0.65V RMS on the 7150 display, the same as my Tek DSO reports. The reading is still drifting around a bit.
>
> * All the resistors in the AC input attenuator and opamp feedback loop are fine.
>
> * At the junction of RLE, C2 and R10 I'm seeing the input signal with a negative-6V DC offset. The output from the opamp is stuck at negative 15V.
>
> This leaves the following suspects:
> Relay RLF (an SDS-Relais RS12 metal can relay)
> IC18 (DG211 analog switch)
> IC15 (LH0062C opamp)
> TR12 (WN1001 FET)
> TR1 and TR2 (3N163 FET)
>
> I have some LH0062H opamps on order -- the tin can variety, not the DIL part. Incidentally the 7150Plus PCB is padded out for the DIL part but a tin-can part has been installed by Solartron and had its leads bent to fit the DIL footprint. Whatever works, I guess.
>
> In summary: I'm getting closer but I'm not quite there yet.
>
> Adjusting the opamp "offset trim" may be quite interesting -- I don't think the adjustment procedure is in the service manual!
>
>
> I'm wondering if it's possible to swap the 3N163s for something else if they prove difficult to locate. A 3-pin FET wouldn't work (it wouldn't allow negative Vds like the 3N163 does)... that means it'd probably have to be an analog switch and some form of level translator.
>
>
> As for the WN1001 - I can't even find a datasheet for it!
> For that matter, the WD460 dual matched FET in the DC loop is also proving difficult to find even a spec sheet.
> I think it's safe to say I'd be pretty well stuffed if the '460 in the DC amp ever failed!
>
>
> I did notice a mildly amusing easter-egg on the schematic: the star-ground on the first floating board schematic is labelled "ROME". As in "All roads lead to Rome"... Reminds me of the "washer lady" and "wizard" who were often found in Tektronix service manuals, back when the schematics were lovingly (?) hand drawn.
>
> Cheers,
> --
> Phil.
> philpem at philpem.me.uk
> http://www.philpem.me.uk/
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.