[volt-nuts] Solartron 7081 calibration "zero too large" error
Fred Schneider
pa4tim at gmail.com
Wed May 2 17:55:59 UTC 2012
Mine does the same, the input in the back was shorted. But maybe I took the wrong connections.
I think I make a little box with banana busses and solder a cable inside instead of the backconnector so I can use it ( on the front instead of jerking the unit of the shelf to reach that ( and if I could find a connector)
Fred PA4TIM
Op 2 mei 2012 om 19:19 heeft Larry Jenkins <lawrence_jenkins at sbcglobal.net> het volgende geschreven:
> Hi I have a Solartron 7061 and during calibration, the unit checks the channel
> (1) inputs on the back of the unit and these must be shorted or you will get
> this error message.
> I went through this same process with my meter. It is not a bug but a feature.
>
> Thank you
>
> Larry
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.