Some further data. Using three different (uncalibrated) DMMs, all three
show exact x10 step increases in voltage for each rotary switch on the
VS330. Although I'm not sure for the right-most switch. I'm in the noise of
the LSD on my best DMM for that one. So, even though I don't know exactly
how accurate this thing is, I feel pretty confident that it is working well.

I've made a few inquiries about getting this calibrated. I'm still waiting
on some answers.

Joe Gray
W5JG



On Sat, Aug 3, 2013 at 4:28 AM, Robert Atkinson <robert8rpi at yahoo.co.uk>wrote:

> Hi Joe,
> To check the differences between the two references, connect the commons
> and put your 3478A meter beween the positive outputs. The meter will read
> the difference. As it should be less than 3V the 3478A will be in high
> impedance mode so loading will be minimised. Reverse the connectons and
> average the two readings to eliminate thermal EMFs. If the VS330 has fine
> enough adjustment you can zero the indicated difference. If the average
> meter reading and the adjustment of the reference for zero match, you have
> a good confidence for the difference between the two standards.
> Now you just need a third standard, maybe a Kelvin Varley divider, another
> meter.............
>
> Robert G8RPI,
>
>
>
>
> ________________________________
>  From: J. L. Trantham <jltran at att.net>
> To: 'Discussion of precise voltage measurement' <volt-nuts at febo.com>
> Sent: Saturday, 3 August 2013, 4:24
> Subject: Re: [volt-nuts] VS 330 arrived
>
>
> Accidents do happen.
>
> No, just kidding.  Probably means all three are very close.
>
> Joe
>
> -----Original Message-----
> From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com] On
> Behalf Of Joseph Gray
> Sent: Friday, August 02, 2013 8:41 PM
> To: Discussion of precise voltage measurement
> Subject: [volt-nuts] VS 330 arrived
>
> The new toy that I don't really need arrived :-) According to my
> uncalibrated HP 3478A, the VS 330 seems to be working fine. The voltages
> are very close to expected values and the increments via the switches are
> consistent.
>
> Not knowing the accuracy of either the DMM or the VS 330, I still thought
> it would be interesting to see what reading I got on the DMM with the VS
> 330 set to 5.00000 VDC.
>
> If you remember, with the MAX6350CPA, the DMM measured 5.0026 VDC @ 26 C.
> Using the VS 330, the DMM reads 5.0020 VDC @ 26 C.
>
> I still need to get both the DMM and the VS 330 calibrated, but it is
> interesting that the only two voltage standards that I have seem to be in
> close agreement.
>
> Joe Gray
> W5JG
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
>


More information about the volt-nuts mailing list