[volt-nuts] Datron 1082

Chuck Harris cfharris at erols.com
Mon Apr 21 11:03:56 EDT 2014


Just in case you haven't yet: Always start any investigation with
checking the power supply for voltage and ripple.  Test the main
electrolytic capacitors for capacitance, and ESR.

Nothing in the unit's self testing ability can be trusted if the
power supply is putting out large amounts of ripple, or is at the
wrong voltage.

-Chuck Harris

Randy Evans wrote:
> I have a Datron 1082 DMM that just recently stopped working.  When I power
> up, it gives a "FAIL" message on the front panel.  I found a flowchart in
> an abbreviated servicing manual that indicates it's from a failure of the
> "Check for uncorrupted calibration store".  I Assume this likely means the
> cal tables are corrupted?  Unfortunately, I don't have sufficient
> information (e.g., schematics) to do any further troubleshooting.  Can
> anyone help with a complete servicing manual and/or further diagnostic help?
>
>
> I did download the Datron 1071DMM  O&S manual from K04BB but I don't know
> how similar it is to the 1082.  Can anyone comment on what the differences
> are?
>
>
>
> Thanks,
>
>
>
> Randy Evans