Tom,
I would think with 0.08 VDC the battery is dead and all the memory is gone.
Do you think it would make any difference if I paralleled the leads with a
new one? I don't have a parts list to know what type the battery is so
it's going to be a challenge to find a direct replacement. Wish I had a
manual.
Randy
On Mon, Apr 21, 2014 at 7:54 PM, Tom Knox <actast at hotmail.com> wrote:
> Randy I would say from my experience with Datron there is about 90% chance
> the battery and recal will solve the problem. I would modify the battery
> with a coupe sets of leads that are long enough so you can install a new
> while the old battery is still in circuit.
> Best Wishes;
> Thomas Knox
>
>
>
> > Date: Mon, 21 Apr 2014 19:23:35 -0700
> > From: randyevans2688 at gmail.com
> > To: volt-nuts at febo.com
> > Subject: Re: [volt-nuts] Datron 1082
> >
> > I believe the voltages are correct but I have no schematics or
> > troubleshooting instructions to confirm. I do see what looks like a
> > battery in the memory section of the 1082 and it measures 0.08 VDC. Not a
> > good sign. I am really hoping someone has a Datron 1082 Service manual
> > with schematics that they can steer me to, preferably a soft copy so I
> can
> > better troubleshoot the unit. It might just need a new battery (plus a
> > re-cal) to get it going again but I hate going blind.
> >
> > Randy
> >
> >
> > On Mon, Apr 21, 2014 at 8:15 AM, Tom Knox <actast at hotmail.com> wrote:
> >
> > > As with many products of this age the tantalum caps are a common point
> of
> > > failure in the 1082.
> > > Although this current problem sounds like a memory battery.
> > >
> > > Thomas Knox
> > >
> > >
> > >
> > > > Date: Mon, 21 Apr 2014 11:03:56 -0400
> > > > From: cfharris at erols.com
> > > > To: volt-nuts at febo.com
> > > > Subject: Re: [volt-nuts] Datron 1082
> > > >
> > > > Just in case you haven't yet: Always start any investigation with
> > > > checking the power supply for voltage and ripple. Test the main
> > > > electrolytic capacitors for capacitance, and ESR.
> > > >
> > > > Nothing in the unit's self testing ability can be trusted if the
> > > > power supply is putting out large amounts of ripple, or is at the
> > > > wrong voltage.
> > > >
> > > > -Chuck Harris
> > > >
> > > > Randy Evans wrote:
> > > > > I have a Datron 1082 DMM that just recently stopped working. When
> I
> > > power
> > > > > up, it gives a "FAIL" message on the front panel. I found a
> flowchart
> > > in
> > > > > an abbreviated servicing manual that indicates it's from a failure
> of
> > > the
> > > > > "Check for uncorrupted calibration store". I Assume this likely
> means
> > > the
> > > > > cal tables are corrupted? Unfortunately, I don't have sufficient
> > > > > information (e.g., schematics) to do any further troubleshooting.
> Can
> > > > > anyone help with a complete servicing manual and/or further
> diagnostic
> > > help?
> > > > >
> > > > >
> > > > > I did download the Datron 1071DMM O&S manual from K04BB but I
> don't
> > > know
> > > > > how similar it is to the 1082. Can anyone comment on what the
> > > differences
> > > > > are?
> > > > >
> > > > >
> > > > >
> > > > > Thanks,
> > > > >
> > > > >
> > > > >
> > > > > Randy Evans
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts at febo.com
> > > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts at febo.com
> > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts at febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
More information about the volt-nuts
mailing list