[volt-nuts] RE "new" 3458A
Randy Evans
randyevans2688 at gmail.com
Sun Aug 17 15:25:58 EDT 2014
Bill,
You are very helpful and I really appreciate it.
I entered BLUE-C-03458 and I still get the ERR message (106, "OUT OF RANGE
-- CAL secured"). That would be consistent with what you are saying. I
guess its no reason to panic yet.
I'm not sure how to use the SECURE command. I guess it's time to open it
up but I will probably have to get some PosiDrive screwdrivers.
Thanks,
Randy
On Sun, Aug 17, 2014 at 11:39 AM, Bill Gold <wpgold3637 at att.net> wrote:
> Randy:
>
> You are doing a ZERO calibration on the meter rather than a ACAL. You
> have to give it a password after the "0" and then it will do the ZERO
> calibration. Usually the password is "3458" and comes from the factory
> that
> way. But someone could have changed it in the past. You can do this from
> the front panel menu using the SECURE command. But there is a jumper
> inside
> the 3458A which disables the request for a password so that you can do any
> CAL or reset the password to what you want. So everything is working as
> expected. The jumper is JM600 on the outguard processor board 66505 or
> 66515 on the left of the meter inside. Try "3458" first as most are set to
> this password. Read the Calibration Manual pg. #8 for how to get inside of
> the meter. You will need two sizes of a PosiDrive type screwdrivers to
> accomplish this. Again in the Cal Manual.
>
> Bill
>
> ----- Original Message -----
> From: "Randy Evans" <randyevans2688 at gmail.com>
> To: "Discussion of precise voltage measurement" <volt-nuts at febo.com>
> Sent: Sunday, August 17, 2014 11:22 AM
> Subject: Re: [volt-nuts] RE "new" 3458A
>
>
> > Bill,
> >
> > I installed the U-short and executed the BLUE-C-0 front panel command and
> I
> > get an ERR on the display. I assume this is not good. Did I do this
> > correctly?
> >
> > Thanks,
> >
> > Randy
> >
> >
> > On Sun, Aug 17, 2014 at 10:49 AM, Bill Gold <wpgold3637 at att.net> wrote:
> >
> > > Randy:
> > >
> > > Look at page #23 of the 3458A Calibration Manual. Make yourself a
> "4
> > > Terminal Short" as shown and put it in as shown.
> > >
> > > Before I did an ACAL I had around -000.00025 mVDC. After ACAL I
> now
> > > read +000.00002 mVDC. I did the "CAL 0" myself so I would expect that
> the
> > > meter should return to a low value, and it does. I do get a variation
> of
> > > +/- 30 nVDC using 100 PLC and just observing the variations. As I
> remember
> > > I have never seen a spec on the ZERO stability over temperature.
> > >
> > > If I turn on the MATH function and then do 40 measurements with 100
> PLC
> > > the statistics show:
> > >
> > > Low reading -70 nVDC
> > > Mean reading -28 nVDC
> > > High reading +3.5 nVDC
> > > Total Variation 73 nVDC
> > >
> > > So that correlates with my visual observation of 60 nVDC. After an
> > > hour
> > > the room had gone up around 1 degree C. Then I observed -000.00023
> mVDC.
> > > After another ACAL the reading was again +000.00002 mVDC. This
> particular
> > > meter has a negative tempco as the room temp goes up.
> > >
> > > Obviously do an ACAL before any precision measurements requiring
> low
> > > nanovolts.
> > >
> > > Go to the Keysight website and go to "Technical Support" and choose
> > > "Parts". Then enter in the "Part Number" "03458-66517" which is the
> > > replacement "03458-66507" assembly and you will see the replacement
> part
> > > number on the right hand side. Click on that and you will get the
> > > information about the exchange program and so on. Looks like you can
> just
> > > order this part online and pay for it with a credit card, but you have
> to
> > > create or use an existing login account.
> > >
> > > I needed a new display a few years ago. At that time you could
> order
> > > just the display for around $80. Being extremely good at removing and
> then
> > > inserting and soldering I ordered the part. The problem was that the
> > > spacing from top to bottom of the pins had changed. It went from
> around
> > > 1.3
> > > inches to around 1.5 inches. So I had to bend the pins to fit my
> display
> > > board and then get something like 72 pins into the holes on the PC
> Board.
> > > It took hours. This change is probably why HP/Agilent/Keysight doesn't
> let
> > > you just get the display anymore but wants you to get the whole PC
> Board
> > > assembly. It did work just fine once installed.
> > >
> > > Hope this helps your decision to keep or not.
> > >
> > > Bill
> > >
> > >
> > >
> > > ----- Original Message -----
> > > From: "Randy Evans" <randyevans2688 at gmail.com>
> > > To: "Discussion of precise voltage measurement" <volt-nuts at febo.com>
> > > Sent: Saturday, August 16, 2014 10:36 PM
> > > Subject: Re: [volt-nuts] RE "new" 3458A
> > >
> > >
> > > > Interesting note. After the room cooled down from about 79F to 73F,
> and
> > > > another ACAL, the meter now reads +000.00035 mVDC, a more reasonable
> > > value,
> > > > although it does bounce around a couple of tenths of a uV.
> > > >
> > > > Maybe that is OK? If so, then the only issue would seem to be the
> > > display
> > > > has some faint pixels, which a new display should fix.
> > > >
> > > > Randy
> > > >
> > > >
> > > > On Sat, Aug 16, 2014 at 9:59 PM, Randy Evans
> <randyevans2688 at gmail.com>
> > > > wrote:
> > > >
> > > > >
> > > > > The unit seems to be working so far except for one issue. After
> doing
> > > an
> > > > > ACAL, and making sure the Auto Zero is ON, I short the input leads
> > > with
> > > a
> > > > > copper wire shunt across the inputs and the reading is
> approximately
> > > > > -000.0023 mVDC. That seems rather high. I would expect the unit
> to
> > > short
> > > > > the input leads internally and force a zero reading during the
> ACAL.
> > > > > Anyone have any comments on this reading?
> > > > >
> > > > > Thanks,
> > > > >
> > > > > Randy
> > > > >
> > > > >
> > > > > On Sat, Aug 16, 2014 at 3:37 PM, Richard Moore
> <richiem5683 at gmail.com>
> > > > > wrote:
> > > > >
> > > > >> Hi Randy -- sounds like your unit is in cal, based on your
> > > measurements
> > > > >> of DCV and precision 10k resistor.
> > > > >>
> > > > >> Using autocal all is recommended before doing precision
> measurements,
> > > and
> > > > >> I do that if it's been more than a day or two since last use. The
> > > autocal
> > > > >> uses the internal Vref and an internal 10K resistor to do cal on
> > > everything
> > > > >> else, so that tells you what the basic cal procedure is. I just
> got
> my
> > > 3458
> > > > >> back from Loveland, and that's what they did for me -- warmed it
> up,
> > > then
> > > > >> ran autocal, then measured everything against a Fluke 5700, aided
> by
> > > an
> > > HP
> > > > >> 3325, and another 3458.
> > > > >>
> > > > >> It has been 5 years since I replaced the display board (no
> "exchange"
> > > > >> deal was available then AFAIK, so I don't know what's changed) and
> > > also
> > > the
> > > > >> NVRAM board, which was dead, with one with the Snap-cap RAM chips.
> I
> > > did
> > > > >> those replacements, then sent it home for cal, which was complete,
> > > since
> > > > >> all the RAM was new. Now after 5 years, the unit passed all
> incoming
> > > > >> performance tests and was sent back to me without a cal process of
> any
> > > > >> kind. This tells me that an old, well-aged Vref module is a good
> > > thing.
> > > The
> > > > >> 10VDC test had changed by a bit under 5ppm, or roughly 1ppm/year.
> > > > >>
> > > > >> They have a cal deal -- use code 1.090 -- press them for it -- and
> > > that
> > > > >> saved me 30% off the normal price. I think this deal lasts until
> > > > >> mid-September, so my recent "cal" ended up at just under $400
> > > including
> > > > >> shipping. I'm not sure the deal is available on new or first-time
> > > cals;
> > > my
> > > > >> unit was in their data bank.
> > > > >>
> > > > >> But this is a long way of saying I don't think you need to send it
> for
> > > > >> cal -- just push Auto Cal and Enter and wait about 10 minutes and
> you
> > > > >> should be good to go.
> > > > >> _______________________________________________
> > > > >> volt-nuts mailing list -- volt-nuts at febo.com
> > > > >> To unsubscribe, go to https://www.febo.com/cgi-bin/
> > > > >> mailman/listinfo/volt-nuts
> > > > >> and follow the instructions there.
> > > > >>
> > > > >
> > > > >
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts at febo.com
> > > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts at febo.com
> > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts at febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
More information about the volt-nuts
mailing list