[volt-nuts] 732A and Prologix received
Randy Evans
randyevans2688 at gmail.com
Sun Aug 24 20:31:14 EDT 2014
Bill,
I am trying to figure out the MATH function without much success. The 3458
On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold <wpgold3637 at att.net> wrote:
> Randy:
>
> The MATH function is accessible from the keypad. I don't have an IEEE
> interface right now that works. You can also program the numeric keypad
> keys to have preprogrammed functions. DEFKEY
>
> I have made my own "low thermal" measurement leads from Pomona #4892
> banana plugs and Belden #9272 wire. Why 9272, because it was handy at the
> time. It is tin plated copper, shielded twisted pair 20 ga. I have plans
> to do custom cables with 16 ga. bare copper wire that I will twist and then
> put a braided shield over it. I simply cannot find what I want so I will
> build my own cable. I have done something like this before and it worked
> fine. When I get a "round toit".
>
> I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
> used
> in the past to verify my homemade "low thermal" leads as described above.
> Frankly I cannot see any difference between using the 1756 cables and my
> homemade cables once I give them a few minutes for the thermals to go away.
> As far as I can tell and measure the differences, if any, are below 0.1 ppm
> at 10 volts.
>
> Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
> adjustable you may be seeing a misadjusted 1 volt from the 732A. As far as
> the instability of the readings it is hard to determine which is causing
> the
> problem. I have programed (DEFKEY) a numeric keypad key #1 with the
> following code. "MATH 14;NRDGS 40;TRIG 4;TRIG ;" So what this does is
> set
> the MATH to "Statistics" (store high reading/low reading/ and mean of the
> readings) in the registers, the number of readings to "40", the trigger to
> "hold" (which keeps the meter from triggering until I press "ENTER" and
> then
> trigger the sequence of 40 readings when I push the "ENTER" button. You
> can
> do all of this manually from the keypads but since I use this sequence a
> lot
> I have preprogrammed it. This is after I set digits to "8" and PLC to 100.
> Once those 40 readings are finished then you can access the various MATH
> statistic registers, using the menu, by entering "MATH" and then a 2 for
> low, a 4 for mean, and 13 for high. Of course you could do all of this
> through the IEEE also. The 3458A has a very rich set of measurement
> commands. I am still learning all of them. It depends upon what I am
> trying to accomplish.
>
> Since the 1.018 and 1.0 volt outputs are passive and derived from
> resistive dividers from the 10 volt, I don't see how they could contribute
> to the varying readings you are measuring. I think I would put a short on
> the input of the 3458A and manually set the range to 1 volt and then
> observe
> the variations that way without the 732A involved. When I do this I see a
> variation from low reading to high reading of 0.125 uVolts and then another
> 40 I get 0.155 uVolts. This is without the GUARD connected to the low side
> of the measurment terminals, GUARD connected doesn't seem to affect the
> readings. So that is the base noise of the 3458A without the 732A,
> somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I got
> a variation of 0.159 uVolts using the same 40 reading method above. I
> would
> use this to determine where your problem might exist. Just having the
> meter
> input shorted will point you in the right direction. Meter, cables or
> 732A.
>
> Sorry for the long dissertation. Friends get mad at me for being so
> detailed sometimes.
>
> Bill
>
> ----- Original Message -----
> From: "Randy Evans" <randyevans2688 at gmail.com>
> To: "Discussion of precise voltage measurement" <volt-nuts at febo.com>
> Sent: Sunday, August 24, 2014 7:22 AM
> Subject: Re: [volt-nuts] 732A and Prologix received
>
>
> > Bill,
> >
> > I have convinced myself that the problem I an seeing is due to thermals.
> > If I move the cables (with gold-plated banana plugs) using a small towel
> > rather than letting my hand touch the plugs, it is much more stable. If
> I
> > then hold the banana plug with my hand after the reading has stabilized,
> > the reading drifts rapidly upward. I am trying to check the stability of
> > the reading but I haven't figured out the MATH function yet. I assume
> this
> > is a programmed function using GPIB only?
> >
> > The stability I am seeing by manually recording readings (using NLPC of
> 100
> > and 1000) is much greater than what you are measuring on your system.
> Not
> > sure how to ascertain if it's the 3458A or the 732. The value of the
> > readings are very different between the two - the 3458 reads about 50 uV
> > high on the 10 V output and about 12 uV low on the 1V output. Rather
> large
> > differences (this is after an ACAL). I need to find some better cables
> to
> > make sure the errors are not due to thermals again.
> >
> > Randy
> >
> >
> > On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold <wpgold3637 at att.net> wrote:
> >
> > > Randy:
> > >
> > > I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
> > > perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
> > > others
> > > that are in this size and package. Order from one of the usual
> electronics
> > > distributors like Allied, Mouser, Digikey. This is a very common
> battery
> > > as
> > > it is used in a lot of "EXIT" signs so they are lighted when the power
> goes
> > > out. I don't see how the 12V 7AH will fit as they are too large. I
> guess
> > > you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
> AH
> > > but the terminals are in the wrong place so you will have to "nibble"
> out
> > > the aluminum plate that holds them in the 732A battery pack. You have
> to
> > > be
> > > careful if you use the 12v 5AH as you will have 4 extra battery
> connection
> > > leads to deal with and connect correctly. I would stick with the 6V
> 4AH.
> > > New batteries will last around 12 to 14 hours before the "CAL" light
> goes
> > > out when AC power is not applied. So shipping to Cal Lab can be a
> problem
> > > if it is a distance away, or you have to use a shipper like UPS or
> FEDEX
> > > and
> > > you ship the night before and then use their "Morning delivery" and the
> Cal
> > > Lab is expecting your 732A. Same on the way back to you. Of course
> you
> > > could always strap another battery on the 732A and hook it up to the
> "ext
> > > power" plug to last longer. I have seen it done. The issue is to get
> the
> > > Cal Lab to charge the extra battery before they ship the 732A back to
> you.
> > >
> > > When you remove and work on the battery pack always have the AC
> power
> > > plugged in. The "CAL" led will stay on because the 18.6 v regulated
> supply
> > > is working.
> > >
> > > The "CAL" light is to indicate that power has not been lost to the
> > > Reference Amp or other associated circuits. When the raw supply
> (battery)
> > > voltage drops below about 21 volts the "CAL" light will go out. Below
> that
> > > voltage the heater circuits will not work correctly and the 18.6 volt
> > > regulated supply will not regulate. The requirement is that the
> Reference
> > > Amp be kept "alive" at all times to maintain the output voltage that
> was
> > > measured at the time of the most recent Calibration or Certification.
> When
> > > the semiconductor junctions are unbiased and cool off when power is
> lost,
> > > and then power is restored the result will be a different 10 volts than
> > > before the power failure. My experience is that after all of the years
> > > that
> > > these units have been powered up, this won't happen and when power is
> lost
> > > and then restored, even months later, the 732A will come back to almost
> > > exactly the same 10 volts as when they lost power, usually with in 0.2
> PPM
> > > after 24 hours of "warm up".
> > >
> > > What type of hookup leads are you using when measuring the 1 volt
> > > output? What is the PLC set to? I always use 100 PLC to measure this.
> If
> > > you don't have "low thermal" connection leads you can experience uV
> changes
> > > for a minute or more after plugging in the leads due to the "thermals"
> > > generated because of the difference in temperature between the banana
> jacks
> > > on the 732A and the banana plugs of the leads. I have found that even
> just
> > > plugging in the lead will generate a thermal difference because of
> > > difference of temps and some heating due to the physical act of just
> > > inserting the plug because of friction between the jack and plug (my
> theory
> > > at any rate). You have to allow at least a minute or more before being
> > > able
> > > to make a measurement after plugging in the leads. I just measured the
> > > variation of the 1 volt output of my 732A and using my 3458A and I got
> a
> > > total difference of 0.159 uV over 40 measurements using 100 PLC on the
> 1
> > > volt range of the 3458A. Using the MATH function and all of the data
> you
> > > can collect. That was after waiting for several minutes after plugging
> in
> > > the leads.
> > >
> > > I hope all of this helps.
> > >
> > > Bill
> > >
> > >
> > >
> > > ----- Original Message -----
> > > From: "Randy Evans" <randyevans2688 at gmail.com>
> > > To: "Discussion of precise voltage measurement" <volt-nuts at febo.com>
> > > Sent: Friday, August 22, 2014 9:03 PM
> > > Subject: Re: [volt-nuts] 732A and Prologix received
> > >
> > >
> > > > Todd,
> > > >
> > > > Thanks for the info. I have several Panasonic 12V 7 AH batteries
> that
> I
> > > > keep topped off and they have very low current draw (~2 to 3 mA at
> 13.5
> > > > VDC) when charged and at their float voltage, so I am pretty sure
> they
> > > are
> > > > in good condition. I will look at getting those in the units after I
> > > > ascertain the condition of the 732.
> > > >
> > > > So now I have a what appears to be a functioning 3458A and a 732A but
> > > they
> > > > slightly disagree. I am like the man with two watches that disagree
> on
> > > the
> > > > time - which is correct? For the moment, i am only concerned with
> > > > stability. The need for absolute accuracy will come later.
> > > >
> > > > Randy
> > > >
> > > >
> > > > On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef <tmicallef at gmail.com>
> > > wrote:
> > > >
> > > > > Randy,
> > > > >
> > > > > You have two possible choices. It can be configured with 4 x 6v 4Ah
> > > > > batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner
> has
> > > > > modified the battery pack already. A couple of mine needed a
> nibbler
> > > tool
> > > > > to remove enough of the aluminum cover that fits over the tops of
> the
> > > > > batteries. The original cover will short out to the battery tabs
> > > regardless
> > > > > of the battery configuration if this is not done.
> > > > >
> > > > > You can find larger capacity batteries that will give you slightly
> more
> > > > > battery life. I lost a couple sets of mail-order batteries after a
> few
> > > > > extended outages. I would recommend going with locally bought
> batteries
> > > > > instead of the cheaper mail order. My local Batteries Plus will
> > > typically
> > > > > have some warranty if I remember correctly. Moving forward I will
> only
> > > use
> > > > > 2 12v batteries and pre-charge them on a battery charger to
> equalize
> > > them
> > > > > before putting them in the 732A. I think the cheap batteries did
> not
> > > > > discharge equally, and would not recover when power was applied.
> > > > >
> > > > > Inspect the back plane for damaged traces and look at the
> capacitors. I
> > > had
> > > > > a few that looked questionable. So far, I have replaced all the big
> > > caps
> > > on
> > > > > the pre-regulator and regulator boards. My feeling is that once
> these
> > > go
> > > > > online, they should run as long as possible between repairs.
> > > > >
> > > > > The battery charger circuit may need adjusting. I tweaked mine and
> it
> > > > > seemed to work fine.
> > > > >
> > > > > Todd
> > > > >
> > > > >
> > > > > On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans <
> > > randyevans2688 at gmail.com>
> > > > > wrote:
> > > > >
> > > > > > I received my Fluke 732A today. Just powered it up but it needs
> new
> > > > > > batteries. Any suggestions for sources (I haven't opened up the
> unit
> > > > > yet -
> > > > > > I want to make sure it works before doing that). Also received
> the
> > > > > > ProLogix USB-GPIB adapter. I plan on using Mark Sims' CAL ran
> data
> > > > > dumper
> > > > > > program to get the CAL data from my 3458A. Should be a busy
> weekend.
> > > > > >
> > > > > > Randy
> > > > > > _______________________________________________
> > > > > > volt-nuts mailing list -- volt-nuts at febo.com
> > > > > > To unsubscribe, go to
> > > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > > and follow the instructions there.
> > > > > >
> > > > > _______________________________________________
> > > > > volt-nuts mailing list -- volt-nuts at febo.com
> > > > > To unsubscribe, go to
> > > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > and follow the instructions there.
> > > > >
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts at febo.com
> > > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts at febo.com
> > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts at febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
More information about the volt-nuts
mailing list