[volt-nuts] Fluke 5100B Calibrator‏ Repair Help

Tom Knox actast at hotmail.com
Thu Mar 13 12:56:03 EDT 2014


Someone will most likely have an exact value, but assuming that RV1 and RV2 are being used as excessive transient voltage protection another way to determine value is to use a variac to power up the 5100B and measure the normal voltages at RV1 and RV2. Then replace with a properly rated MOV. 

Thomas Knox



> From: w6te at msn.com
> To: volt-nuts at febo.com
> Date: Thu, 13 Mar 2014 09:34:48 -0700
> Subject: [volt-nuts] Fluke 5100B Calibrator‏ Repair Help
> 
> I need some help with repairing a Fluke 5100B Calibrator. 
> The capacitor support bracket fell into the power supply module and vaporized both varistors, RV1 and RV2. A call to Fluke tells me they no longer can supply these MOV's. According to the manual's parts list these varistors are in-house parts by Fluke. I have no way of determining a substitute part. Is there anyone that can help me identify what could be a useful replacement?
> Thanks,
> Dave - W6TE 		 	   		  
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
 		 	   		  