[volt-nuts] 3458a RAM batteries - longer life if unit is continually powered?
Tom Knox
actast at hotmail.com
Thu Mar 27 13:59:02 EDT 2014
Date Codes can be misleading because most of this style of chip does not connect to the battery until the first power up cycle. So evl
Thomas Knox
> Date: Thu, 27 Mar 2014 17:33:17 +0100
> From: frank.stellmach at freenet.de
> To: volt-nuts at febo.com
> Subject: [volt-nuts] 3458a RAM batteries - longer life if unit is continually powered?
>
> Hi,
>
> the DALLAS datasheet specifies 10 years data retention. Quite obviously,
> that's the unpowered time.
>
> That time includes self discharge and RAM sustaining currents.
>
> self discharge and RAM supply current will be of the same order, i.e. <
> 1µA, I assume from experience.
> And I assume also, that those currents will vary over samples, i.e. over
> samples of battery and RAM.
>
> Recently, there was the case, where the CAL RAM was still alive after 24
> years (instrument vintage 1989, same DALLAS date code!), but with
> unknown powering history.
>
> Additionally, a typical 10 years of lifetime is given also for those
> batteries.
>
> Therefore, in practise there will be no pronounced difference between
> powered / unpowered state.
> On the other hand, DALLAS implemented an electronic seal, so that the
> RAM inside the package is initially powered only after first power up.
> But there's no parameter in the datasheet, specifying data retention
> according to date code or to initial power up.
>
>
> The waste of energy just to extend the batteries lifetime, does not pay off.
> On the contrary, the MTBF time of the instrument will be consumed, the
> display will get bad, and here in Germany, this energy consumption would
> cost a fortune (meanwhile: 25+ Cents / kWh, crazy).
>
>
> Yes, as long the instrument is powered, the CAL RAM will not lose its
> content.. but what about unexpected power outages?
>
>
> If you save the CAL RAM and repogram a fresh one after failure, there's
> no need to send it to agilent.
>
>
> No, not any RAM error needs recalibration.
> Those error messages are not documented.
>
> So I've scanned the text fragments in the firmware file:
>
> RAM TEST 1
> RAM TEST 2
>
> probably refer to either to the configuration RAMs (2x32kB), andthe CAL
> RAM (8kB).
>
> Eror messages are:
> NONVOLATILE RAM HiGH /LOW
> CAL RAM FAILURE
>
> and known ones from other users:
>
> RAM TEST 1 HIGH / LOW
>
> So I assume, that a real failure of the cal constants only apply on
> message: CAL RAM FAILURE and perhaps RAM TEST 2 ...., as RAM TEST 1
> should refer to the configuration RAMs only, HIGH to the high byte RAM,
> LOW to the low byte RAM.
>
> Errors of the cal RAM will manifest in additional ACAL or CAL failures.
>
>
> Frank
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
More information about the volt-nuts
mailing list