[volt-nuts] Calibration of voltage standards

wb6bnq wb6bnq at cox.net
Wed Feb 11 02:55:05 EST 2015


To amplify Chuck's point,

The only meter worth considering, if you are going to get one, is the 
FLUKE 845.  Preferably, the battery operated version which has a higher 
isolation leakage resistance.  However, the 110
volt rack mount model would do just as equally.

Bill....WB6BNQ


Chuck Harris wrote:

> To do a comparison of the sort you are asking about, the
> sensitivity of the null meter is much more important than
> its ultimate accuracy.
>
> So, neither of your meters is really the right meter to
> use for this task.  What you want is called a null meter,
> and is generally sensitive to the microvolt region.
>
> -Chuck Harris
>
> Ken Peek wrote:
>
>> Hi Group,
>>
>> I have heard of a few different ways to measure one 10V voltage standard
>> against another 10V voltage standard.
>>
>> Assume we have two 10V voltage standards.  One is calibrated, the other
>> not only needs to be calibrated, but probably adjusted.  For the sake of
>> simplicity, let's say the two standards are Fluke 732B's.
>>
>> I *think* the best way is to connect the two units' (-) terminals
>> together, then connect a calibrated meter in between the (+) terminals,
>> and measure the difference.  I have also heard that to remove thermal
>> EMFs, you should use a low-thermal-EMF DPDT switch or a low-thermal-EMF
>> relay to reverse the connections on the DMM, so you can take the reading
>> forward and reversed, then split the difference.  There is the
>> possibility to introduce thermal-EMF errors from the switch/relay as
>> well, so I'm wondering if this is a good idea.  This sort of makes sense
>> to me, but I'm not a metrologist, so I would like to hear what others in
>> this group think about this.
>>
>> So, just what is the proper way to accomplish this task?
>>
>> BTW-- I have an Agilent 34420A and an HP 3458A, which would be the
>> better instrument for this task?
>>
>> Best Regards,
>> Ken Peek
>> =============================
>>
>>
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts at febo.com
>> To unsubscribe, go to 
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to 
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>



More information about the volt-nuts mailing list