I bought two HP-3458A ref boards from John Daly (he no longer has any more)
and they worked fine but have intermittent noise bursts.  I suspect they
were rejects from HP/Agilent.  At some point I will probably replace the
LTZ1000As and reduce the internal temperature set point for better
stability.  Has anyone else had any issues with these units?

BTW, I packaged the units in an enclosure for a stand-alone reference and,
except for the occasional noise bursts, seem very stable according to my HP
3458A and fluke 732A/ESI RV722 KVD.

Randy

On Wed, Jan 28, 2015 at 9:40 AM, John Phillips <john.phillips0 at gmail.com>
wrote:

> There is a bottom cover under the board.
>
> On Tue, Jan 27, 2015 at 9:01 PM, J. L. Trantham <jltran at att.net> wrote:
>
> > When these boards are installed in a 3458A, the LTZ1000 is covered by a
> > small, white, 'hat' that mounts with screws that pass through the two
> holes
> > on opposite sides of the LTZ1000.  I've never removed the board to see if
> > there is anything on the bottom of the board that would cover the bottom
> of
> > the LTZ1000.
> >
> > Does anyone know what this 'hat' is constructed of or the
> > HP/Agilent/Keysight part number of this item is?  Is it available from
> > Keysight?
> >
> > In the 3458A the reference board is under a cover that covers about half
> > of the top of the meter but not 'insulated' beyond that, except for the
> > 'hat' referred to above.
> >
> > It would appear desirable to find what that 'hat' is and see if something
> > could be fabricated to 'insulate' the LTZ1000 for a 'do it yourself'
> > standalone reference.
> >
> > Would appreciate hearing anything about how the reference board is
> > 'insulated' in the 3458A.
> >
> > Thanks.
> >
> > Joe
> >
> >
> > -----Original Message-----
> > From: volt-nuts [mailto:volt-nuts-bounces at febo.com] On Behalf Of Orin
> Eman
> > Sent: Tuesday, January 27, 2015 11:45 AM
> > To: Discussion of precise voltage measurement
> > Subject: [Bulk] Re: [volt-nuts] 3458A reference boards on ebay
> >
> > Dave,
> >
> > I'll be putting it in an enclosure to use as a reference.  It will
> > probably be a smallish diecast aluminium box.  It's what I did with my
> > Geller SVR-T and it compared very well with my Fluke 731B.  I'll likely
> use
> > Pomona 3770 binding posts for output (I used them to repair the 731B).
> > I'll also use the protection circuits from the Fluke 732 on the output -
> > big reverse diode, gas discharge tube and TVS as I recall (I plan to add
> > these protection circuits to the 731B).
> >
> > I'll likely do a PCB for it, though the local equivalent of veroboard
> > would likely be OK; the socket spacing is 0.9" and the pin spacing 0.1".
> > My meters are also 6 1/2 digit (3456A and 34461A) so I'm not too
> concerned
> > with thermal EMFs with one count being 10uV on the 10V range.  However, I
> > will try to avoid temperature gradients and keep the outputs close
> > together.  I'll use star wiring for the 18V and ground as in the
> following:
> >
> > http://www.maxmcarter.com/vref/
> >
> > It is debatable whether the -15V supply is needed - I'll test with and
> > without.  I'd add a few tantalum capacitors to the above design - on the
> > regulator outputs and on the 3320 resistor, along with protection diodes
> > for the LM317.
> >
> > There is also a HUGE thread about the LTZ1000 on EEVBlog - I'll be
> > reviewing that for ideas:
> >
> > http://www.eevblog.com/forum/projects/ultra-precision-reference-ltz1000/
> >
> > I actually have a couple of LTZ1000As here that I was going to DIY, but
> > never got around to getting the precision resisters .  I think there is a
> > source in England mentioned in the EEVBlog thread.
> >
> > Orin.
> >
> > On Tue, Jan 27, 2015 at 2:24 AM, Dr. David Kirkby (Kirkby Microwave Ltd)
> <
> > drkirkby at kirkbymicrowave.co.uk> wrote:
> >
> > > On 27 January 2015 at 09:16, Orin Eman <orin.eman at gmail.com> wrote:
> > > > There is a seller letting a slow trickle of them out there... they
> > > > seem
> > > to
> > > > be going for about $165.  The current batch of two is around $100,
> > > > but
> > > will
> > > > probably be bid up.
> > > >
> > > > I just received one that I won.  Seems to be working fine after a
> > > > quick breadboard lashup.  I'll be making an enclosure for it next.
> > > >
> > > > A little expensive IMO, but given the trouble getting the precision
> > > > resistors to DIY with the LTZ1000A, probably worth it.
> > > >
> > > > Orin
> > >
> > > Orin,
> > >
> > > What do you intend doing with it?
> > >
> > > I was thinking about getting one of those, with a view to putting it
> > > in a box with a couple of terminals to have something to compare with
> > > my 6.5 digit 3457A.  But what put me off is a lack of knowledge in
> > > knowing how to convert a 3458A reference board into a boxed unit with
> > > a known output voltage at the terminals. How would I avoid / control
> > > thermal EMFs?
> > >
> > > I'd be interested to hear what your plans are for it, and how you
> > > intend tackling those issues.
> > >
> > > Do you know what the difference in the reference is between a standard
> > > 3458A (8 ppm) and the high stability option 002 (4 ppm) model? I'm
> > > guessing the chips for the option 002 might be the top performing
> > > ones. I wonder if there's any way to tell from your board if it came
> > > from a standard 3458A or a 3458A with option 002.
> > >
> > > Dave
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts at febo.com To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts at febo.com To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts at febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> >
>
>
>
> --
>
> *John Phillips*
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>


More information about the volt-nuts mailing list