[volt-nuts] 3458A Zero reading
Randy Evans
randyevans2688 at gmail.com
Mon Jul 27 20:49:28 EDT 2015
I used a 12 AWG wire formed into a U shape for the short and the zero
voltage measured less than 0.2 uV after a good 2+ hour warmup. However, I
am seeing drifts of several uV for the 752 calibration. I was using TV
twin lead to connect the 732 to 752 and 752 to 3458A since that's really
all I have for now. I suspect thermals are killing me so I need to find
some low thermal shielded differential cable to connect from the 752 to the
3458A. I did not see anything like that in Pomona's arsenal. I was
thinking about Bill Gold's idea of making my own but I think it would end
up being \very stiff. Any other ideas for a good source of shielded copper
wire twinax?
Randy AE6YG
On Mon, Jul 27, 2015 at 4:10 PM, J M via volt-nuts <volt-nuts at febo.com>
wrote:
> I use a 8 oz styrofoam cup to isolate air currents around the 3458A
> terminals when zeroing with shorting bar
>
>
> Sent from my 20" Texas sized
> iPhone....
>
>
> > On Jul 27, 2015, at 2:57 PM, Richard Moore <richiem5683 at gmail.com>
> wrote:
> >
> >
> > Randy — My 3458A manual specs 12 or 14 gauge solid copper wire for the
> short, bunt in to a U shape. I use 14 gauge. Poul-Henning’s comments about
> thermals are well-taken — even a simple folded up carboard shield helps.
> But I think you’re near the limit.
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts at febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
More information about the volt-nuts
mailing list