The Fluke 5700A and 5720A as well as the Wavetek 4808 and 4800A are pretty amazing I have seen several that appear to have been abused for years without a Cal and are still spot on. Same with many of the HP and Agilent 3458A's.
It seems I see more broken then out of spec.
Thomas Knox
> From: tmicallef at gmail.com
> Date: Fri, 15 May 2015 15:17:19 -0400
> To: info at electronicsandbooks.com; electronicsandbooks at yahoo.com; volt-nuts at febo.com
> Subject: Re: [volt-nuts] How accurate is a fluke after 9 years
>
> Frans,
>
> Do you perform an in-house calibration with a 10V reference and 1 ohm & 10k resistors? I have been studying the recommended calibration intervals of Fluke's 57xx series calibrators and I wonder how well the unit maintains its stability over long intervals between trips to a cal lab.
>
> Also, if you don't mind, would you give a ballpark figure for the cost of cal from your cal lab?
>
> Todd
>
> Sent from my iPad
>
> > On May 15, 2015, at 14:50, Electronics and Books via volt-nuts <volt-nuts at febo.com> wrote:
> >
> > My Fluke 5700A calibrator was calibrated today after 9 years. It displays how much it deviated from the last calibration. For the both voltage references i got 0.0 ppm and for the internal 10k resistor -1.7 ppm.
> > Only the amplifier for the 11-22 volt range drifted 1700 ppm for it depends on a voltage divider composed of two normal metalfilm resistors. These will be replaced by 2 vishay 0.01% resistors which are hopefully more stable.
> >
> > Regards
> >
> > Frans
> >
> >
> >
> > Discere ne cesses
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts at febo.com
> > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
More information about the volt-nuts
mailing list