[volt-nuts] volt-nuts Digest, Vol 80, Issue 14; Solratron cal issue
David C. Partridge
david.partridge at perdrix.co.uk
Sat Apr 23 08:37:36 EDT 2016
Yes, there is:
The monitor mode command "I" (upper case i) will write a test pattern to the NVM and checksum it.
************** Any calibration constants will be lost **********************
As you don't have the manuals handy here's the long form:
Use your favourite software terminal emulator (e.g. Putty) to connect to the 7081 using the serial port. (straight through serial cable).
The baud rate is determined by dip-switch S1 on the earthy logic board. If all serial port rate switches are to the right, it's set to 9600 baud.
Issue the following commands:
out,rs232,on
monitor
I
Ctrl-X
Dave
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces at febo.com] On Behalf Of Four Designs Company
Sent: 22 April 2016 17:20
To: volt-nuts at febo.com
Subject: Re: [volt-nuts] volt-nuts Digest, Vol 80, Issue 14; Solratron cal issue
It has been some years since I last worked on a 7081, and I don't have the manuals at hand..
Is there a way to clear the EAROM contents and perhaps load starting cal constants? I think the EAROM data is corrupt or so far off that it cannot compute new values.
Bob
More information about the volt-nuts
mailing list