[volt-nuts] Guidance requested.
Charles Steinmetz
csteinmetz at yandex.com
Mon Jul 10 06:07:05 EDT 2017
David wrote:
> So please how best to proceed from here?
The Fluke 732A is by far the most accurate and most stable voltage
reference you have. There are only a few cal labs worldwide that can
calibrate it to an uncertainty that does it justice (1/10 ppm). Most
good to excellent cal labs have 10v DC uncertainties from 1-10 ppm.
SO: First, have the 732A calibrated by one of those few labs. I
strongly favor the Fluke US cal facility (Fluke Park Laboratory,
Everett, WA, USA, NVLAP Lab Code 105016-0), with an accredited DC
calibration uncertainty of 100nV at 10v. However, it may be an issue
getting it back to you still under power from there (IIRC, you are in
the UK). I maintain three 732As and calibrate each one annually on a
rotating schedule, one every 4 months. I built a shipping container
with internal batteries that will keep the 732A powered for three days,
if necessary, and an internal charger so all Fluke has to do is plug the
shipping container in to recharge them.
If the choice comes down to a lab that can calibrate the 732A with full
accuracy (1/10 ppm), but it will arrive back to you "cold," and a lab
that cannot calibrate it to that standard in the first place, I'd choose
the former. In my experience, 732As generally retrace within 0.25 ppm
if re-powered within a few days. IOW, I recommend the Fluke US cal lab
unless you have a trustworthy cal lab with 1/10 ppm uncertainty local to
you. (But although experience suggests that you can rely on the
re-powered standard to be within 0.25 ppm of the calibrated value, you
will have lost NIST traceability.)
So whatever you do, have the 732A calibrated *properly* before you do
anything else.
Also, put the 732A on its own dedicated 2-3 kW "online"-type UPS. If
the UPS runs nothing but the 732A, it should carry you through all but
the longest power outages. Adding external batteries is a good plan
(industrial "online" UPSs support this). I have a backup generator, but
even so my UPS system will run the 732As, my primary frequency
references, and transfer DMM for more than a week if necessary.
Note that "calibrating" the 10v output of a 732A generally does not
involve adjusting it. Rather, the cal lab will tell you its actual
voltage to 1/10 ppm, and you will calculate from that value when you use
it. I usually have a new-to-me standard adjusted the first time I have
it calibrated, unless I obtained it with its full calibration history
from an unimpeachable source. I obtained two of my 732As this way, one
I repaired. I had that one adjusted at its first two calibrations.
Finally, note that proper calibration of a 732A is not cheap. Mine are
on a calibration contract, and I still pay over $1000 (plus overnight
shipping two ways on 40+ pounds). That said, the real value of a 732A
is its calibration history. Without a proper calibration and the
calibration history, it is just another voltage reference that needs
characterization before it can be relied on.
Best regards,
Charles
More information about the volt-nuts
mailing list