[volt-nuts] 3458A question
David C. Partridge
david.partridge at perdrix.co.uk
Thu Jan 11 08:56:00 EST 2018
To add some more information
The meter has now been on for a number of days with a newly cleaned 4 wire short in place all the time.
I have run a full CAL 0 to zero the offsets, but even so, and after running an ACAL 0, I get a DCV reading of -200 to -400nV (0.00040mV).
Is that normal? I'd have thought it would typically be less than that given that the 1 year limit is 0.00106mV (1.06Uv).
Thanks again
David
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces at febo.com] On Behalf Of David C. Partridge
Sent: 10 January 2018 10:25
To: 'Discussion of precise voltage measurement'
Subject: Re: [volt-nuts] 3458A question
Correction to -0.00060mV to -0.00070mV
Dave
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces at febo.com] On Behalf Of David C. Partridge
Sent: 10 January 2018 10:14
To: 'Discussion of precise voltage measurement'
Subject: [volt-nuts] 3458A question
After running an ACAL 0 with a four wire short in place I get a DCV reading starting at about -0.00005 mV to increasing to about 0.00060 to 0.00070 mV over a minute or so at which point it seems to stabilise.
Is this "normal" and if not what could be wrong (AZERO is turned on).
Dave
_______________________________________________
volt-nuts mailing list -- volt-nuts at febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
_______________________________________________
volt-nuts mailing list -- volt-nuts at febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
More information about the volt-nuts
mailing list