[volt-nuts] What's the probability of a random used 3458A passing a Keysight calibration?
David C. Partridge
david.partridge at perdrix.co.uk
Tue Jan 16 17:52:10 EST 2018
Surely the AC errors if any must be associated with the A2 AC Convertor board - 03458-66502?
Dave
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces at febo.com] On Behalf Of Dr. Frank
Sent: 16 January 2018 22:39
To: volt-nuts at febo.com
Subject: [volt-nuts] What's the probability of a random used 3458A passing a Keysight calibration?
I also saw this, so at least, they will have adjusted this range.
Yes you're right, the ADC is the A3 board. I confused that with the A5 controller board, which is the only one which had been redesigned to SMT components.
There's no chance to identify a refurbished or replaced board, other than opening the case and looking inside.
And even then, it's not evident, that the ADC IC is ok. You only might do the test according to AN-18.
Replacement boards will have the same part number, which changes only by redesign.
Frank
_______________________________________________
volt-nuts mailing list -- volt-nuts at febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
More information about the volt-nuts
mailing list