[volt-nuts] Prema 6048 zero offset problems

David C. Partridge david.partridge at perdrix.co.uk
Fri Jan 19 13:00:06 EST 2018


If there are any other members of the group with a 6048, I'd be grateful if you would attempt to recreate the scenario I describe to determine if this is just my meter or impacts other 6048 users.

Thanks
Dave

-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces at febo.com] On Behalf Of David C. Partridge
Sent: 18 January 2018 11:57
To: 'Discussion of precise voltage measurement'
Subject: [volt-nuts] Prema 6048 zero offset problems

Starting with a warmed up Prema 6048, if I power it off and then on, select 20V DC range, 4 second integration and connect to the 10V output of a Fluke 732A, I get a reading of:

 +10.000047

If I reverse the polarity of the cables and allow the reading to settle down for a minute, I get a reading of:

 -10.000006

If I short the inputs of the meter, I get a reading of:

 +0.000105

which is different from the  +0.000020 which I would have expected from the readings of the 732A output.

Proceeding on the assumption that the meter has stored a zero offset of -105uV

I then press the ZERO button to null out the offset and then reconnect the 732A in normal polarity I get an incorrect reading of:

 +10.000153

This is not what I would expect to see - I would expect to see a reading of
+10.000047 - 0.000105 = +09.999942  This calculation

I then reverse the inputs and get a reading of:

 -10.000113

which is close to the expected reading of -10.000006 - 0.000105 = -10.000111

I have mailed Prema about this problem, and will let you know what they have to say.

Cheers
David Partridge

_______________________________________________
volt-nuts mailing list -- volt-nuts at febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.



More information about the volt-nuts mailing list