Update: I reseated connectors and got the 1 MHz and 100 kHz divider chains working.

I found that the stiffness in the fine tuning mechanism was caused by a binding turns counting dial. The rest of the mechanism moved freely and to both extremes of rotation when the dial was removed. I disassembled, cleaned, and lubed the dial, and then reinstalled it. Now everything works as it should. I'm letting the standard stabilize before I attempt any adjustments.

One comment about the fine frequency adjustment. One side of the trimmer capacitor is the copper rod, and the other side is a metal cylinder that has insulation (dielectric?) on the inner surface and is held stationary while the rod advances into or retracts out of it, thus changing the value of the trimming capacitance.

What stood out to me when I examined this tuning capacitor mechanism is that the cylinder is attached to a short length of stiff wire that is then soldered to a nearby contact on the PC board. The only thing holding this cylinder in position is the stiffness of the wire. Since we're talking about being able to adjust the frequency in 5x10E-11 increments, with a total range of 1x10E-7 given by a travel of the rod of less than 1/2", I would think that any impact, vibration or other mechanical stress might cause the cylinder to shift, thereby changing the trimmer capacitor value. Maybe this standard is only supposed to be used in a vibration-free and stress-free environment, but it still seems like a poor design in an otherwise well-engineered frequency standard.

Thanks again to everyone for their help with this.

Stan

Update: I reseated connectors and got the 1 MHz and 100 kHz divider chains working. I found that the stiffness in the fine tuning mechanism was caused by a binding turns counting dial. The rest of the mechanism moved freely and to both extremes of rotation when the dial was removed. I disassembled, cleaned, and lubed the dial, and then reinstalled it. Now everything works as it should. I'm letting the standard stabilize before I attempt any adjustments. One comment about the fine frequency adjustment. One side of the trimmer capacitor is the copper rod, and the other side is a metal cylinder that has insulation (dielectric?) on the inner surface and is held stationary while the rod advances into or retracts out of it, thus changing the value of the trimming capacitance. What stood out to me when I examined this tuning capacitor mechanism is that the cylinder is attached to a short length of stiff wire that is then soldered to a nearby contact on the PC board. The only thing holding this cylinder in position is the stiffness of the wire. Since we're talking about being able to adjust the frequency in 5x10E-11 increments, with a total range of 1x10E-7 given by a travel of the rod of less than 1/2", I would think that any impact, vibration or other mechanical stress might cause the cylinder to shift, thereby changing the trimmer capacitor value. Maybe this standard is only supposed to be used in a vibration-free and stress-free environment, but it still seems like a poor design in an otherwise well-engineered frequency standard. Thanks again to everyone for their help with this. Stan
BC
Bob Camp
Wed, Mar 20, 2024 1:07 PM

Hi

These standards (and similar ones from various other outfits) were designed for rack mount / stationary use. Things like glass dewar’s for insulation are a fine thing if you are stationary. They are not all that great if you undergo a lot of shock and vibration. ( Also not great if somebody is shipping one of those designs to you via UPS or FedEx …). There are a number of “not so rugged” items like this in the typical designs ….

As a side note: they also “assumed” that power would be applied continuously 24 hours a day, 7 days a week. Fast warmup and fast stabilization …. not a major design goal.

Bob

On Mar 19, 2024, at 9:52 PM, Stan via time-nuts time-nuts@lists.febo.com wrote:

Update: I reseated connectors and got the 1 MHz and 100 kHz divider chains working.

I found that the stiffness in the fine tuning mechanism was caused by a binding turns counting dial. The rest of the mechanism moved freely and to both extremes of rotation when the dial was removed. I disassembled, cleaned, and lubed the dial, and then reinstalled it. Now everything works as it should. I'm letting the standard stabilize before I attempt any adjustments.

One comment about the fine frequency adjustment. One side of the trimmer capacitor is the copper rod, and the other side is a metal cylinder that has insulation (dielectric?) on the inner surface and is held stationary while the rod advances into or retracts out of it, thus changing the value of the trimming capacitance.

What stood out to me when I examined this tuning capacitor mechanism is that the cylinder is attached to a short length of stiff wire that is then soldered to a nearby contact on the PC board. The only thing holding this cylinder in position is the stiffness of the wire. Since we're talking about being able to adjust the frequency in 5x10E-11 increments, with a total range of 1x10E-7 given by a travel of the rod of less than 1/2", I would think that any impact, vibration or other mechanical stress might cause the cylinder to shift, thereby changing the trimmer capacitor value. Maybe this standard is only supposed to be used in a vibration-free and stress-free environment, but it still seems like a poor design in an otherwise well-engineered frequency standard.

Thanks again to everyone for their help with this.

Stan


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

Hi These standards (and similar ones from various other outfits) were designed for rack mount / stationary use. Things like glass dewar’s for insulation are a fine thing if you are stationary. They are not all that great if you undergo a lot of shock and vibration. ( Also not great if somebody is shipping one of those designs to you via UPS or FedEx …). There are a number of “not so rugged” items like this in the typical designs …. As a side note: they also “assumed” that power would be applied continuously 24 hours a day, 7 days a week. Fast warmup and fast stabilization …. not a major design goal. Bob > On Mar 19, 2024, at 9:52 PM, Stan via time-nuts <time-nuts@lists.febo.com> wrote: > > Update: I reseated connectors and got the 1 MHz and 100 kHz divider chains working. > > I found that the stiffness in the fine tuning mechanism was caused by a binding turns counting dial. The rest of the mechanism moved freely and to both extremes of rotation when the dial was removed. I disassembled, cleaned, and lubed the dial, and then reinstalled it. Now everything works as it should. I'm letting the standard stabilize before I attempt any adjustments. > > One comment about the fine frequency adjustment. One side of the trimmer capacitor is the copper rod, and the other side is a metal cylinder that has insulation (dielectric?) on the inner surface and is held stationary while the rod advances into or retracts out of it, thus changing the value of the trimming capacitance. > > What stood out to me when I examined this tuning capacitor mechanism is that the cylinder is attached to a short length of stiff wire that is then soldered to a nearby contact on the PC board. The only thing holding this cylinder in position is the stiffness of the wire. Since we're talking about being able to adjust the frequency in 5x10E-11 increments, with a total range of 1x10E-7 given by a travel of the rod of less than 1/2", I would think that any impact, vibration or other mechanical stress might cause the cylinder to shift, thereby changing the trimmer capacitor value. Maybe this standard is only supposed to be used in a vibration-free and stress-free environment, but it still seems like a poor design in an otherwise well-engineered frequency standard. > > Thanks again to everyone for their help with this. > > Stan > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com
PS
paul swed
Wed, Mar 20, 2024 1:14 PM

I don't know that there was ever any restriction on the use of the sulzers.
But I think they were the late 1950s and 60s and maybe they did not have
the glass variable caps back then. Sort of doubt that. By the way you are
correct on the capacitors design.
It does seem odd. But you have it working, be happy.
Paul
WB8TSL

On Tue, Mar 19, 2024 at 11:31 PM Stan via time-nuts <
time-nuts@lists.febo.com> wrote:

Update: I reseated connectors and got the 1 MHz and 100 kHz divider chains
working.

I found that the stiffness in the fine tuning mechanism was caused by a
binding turns counting dial. The rest of the mechanism moved freely and to
both extremes of rotation when the dial was removed. I disassembled,
cleaned, and lubed the dial, and then reinstalled it. Now everything works
as it should. I'm letting the standard stabilize before I attempt any
adjustments.

One comment about the fine frequency adjustment. One side of the trimmer
capacitor is the copper rod, and the other side is a metal cylinder that
has insulation (dielectric?) on the inner surface and is held stationary
while the rod advances into or retracts out of it, thus changing the value
of the trimming capacitance.

What stood out to me when I examined this tuning capacitor mechanism is
that the cylinder is attached to a short length of stiff wire that is then
soldered to a nearby contact on the PC board. The only thing holding this
cylinder in position is the stiffness of the wire. Since we're talking
about being able to adjust the frequency in 5x10E-11 increments, with a
total range of 1x10E-7 given by a travel of the rod of less than 1/2", I
would think that any impact, vibration or other mechanical stress might
cause the cylinder to shift, thereby changing the trimmer capacitor value.
Maybe this standard is only supposed to be used in a vibration-free and
stress-free environment, but it still seems like a poor design in an
otherwise well-engineered frequency standard.

Thanks again to everyone for their help with this.

Stan


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

I don't know that there was ever any restriction on the use of the sulzers. But I think they were the late 1950s and 60s and maybe they did not have the glass variable caps back then. Sort of doubt that. By the way you are correct on the capacitors design. It does seem odd. But you have it working, be happy. Paul WB8TSL On Tue, Mar 19, 2024 at 11:31 PM Stan via time-nuts < time-nuts@lists.febo.com> wrote: > Update: I reseated connectors and got the 1 MHz and 100 kHz divider chains > working. > > I found that the stiffness in the fine tuning mechanism was caused by a > binding turns counting dial. The rest of the mechanism moved freely and to > both extremes of rotation when the dial was removed. I disassembled, > cleaned, and lubed the dial, and then reinstalled it. Now everything works > as it should. I'm letting the standard stabilize before I attempt any > adjustments. > > One comment about the fine frequency adjustment. One side of the trimmer > capacitor is the copper rod, and the other side is a metal cylinder that > has insulation (dielectric?) on the inner surface and is held stationary > while the rod advances into or retracts out of it, thus changing the value > of the trimming capacitance. > > What stood out to me when I examined this tuning capacitor mechanism is > that the cylinder is attached to a short length of stiff wire that is then > soldered to a nearby contact on the PC board. The only thing holding this > cylinder in position is the stiffness of the wire. Since we're talking > about being able to adjust the frequency in 5x10E-11 increments, with a > total range of 1x10E-7 given by a travel of the rod of less than 1/2", I > would think that any impact, vibration or other mechanical stress might > cause the cylinder to shift, thereby changing the trimmer capacitor value. > Maybe this standard is only supposed to be used in a vibration-free and > stress-free environment, but it still seems like a poor design in an > otherwise well-engineered frequency standard. > > Thanks again to everyone for their help with this. > > Stan > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com >